Author Topic: Measuring nanoamps and below like a Ninja  (Read 117460 times)
Offline BFX
  • Country: sk
Re: Measuring nanoamps and below like a Ninja
« Reply #75 on: February 19, 2017, 07:32:41 pm »
A calibration box for a hi-pot tester?
http://www.sefelec.com/en/calibration-kit-MG-91
We will see how usable it is  :D   :-DMM
Offline guenthert
« Reply #76 on: February 19, 2017, 11:18:26 pm »
I like the idea of coiling the leads of the resistors.  Might have prevented my mishap, breaking the glass of a 100GOhm resistor.
Offline ocw
« Reply #77 on: March 04, 2017, 12:47:38 am »
There has been slow evolution of my ADA4530-1 low current to voltage converter.  It has selectable 1k, 10k, 100k, 1M, 10M, 100M, 1G, 10G, 100G and 1T feedback resistors.  See its attached picture.
I'm trying to find the minimum current which I can measure with the 1T feedback resistor.  I knew that 1.5 pA was easy by using a 1.55 volt battery which was connected to an external 1T resistor.  I added a 1/10 voltage divider inside of my shielded test box to feed 155 mV to that resistor.  The second attachment shows the 165 fA current produced by that combination (1T 10% resistor is actually close to 0.955).  As is typical with these converters, a positive current is converted to a negative voltage.  1 volt = 1 pA = 1,000 fA.
My test box has a switch to change to a 1/100 voltage divider.  The third attachment shows the approximately 16 fA current produced by that.
Time to build a greater voltage divider.
Offline VintageNutTopic starter
« Reply #78 on: March 04, 2017, 02:17:14 am »
Two things that I am surprised about
1. No guard
2. The banana jacks appear to garden variety. These usually have significant leakage compared to pA current
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
Offline ocw
« Reply #79 on: March 04, 2017, 02:34:32 am »
On the ADA4530-1 data sheets Analog Devices warns of possible nonlinearity when using high voltage resistors at low voltages (a poor voltage coefficient).  To try and evaluate the linearity of the 1T ohm resistor which I used as both a feedback resistor, with another used along with a metered voltage to set the current level which is being measured, I fed the external MOX112523100AK resistor with a 5 volt ramp waveform and observed the output of the converter.
I fed a 3 milliHertz waveform to the converter's input.  While my DC measurements with the battery and resistors located inside of a shielded box produce negligible power line noise, my quick connection to the function generator had some of that noise modulating the 3 milliHetz source.  But, the attached view of the output of the converter still looks like the MOX112523100AK resistor is fairly linear over the 0 - 5 volt range measured.
Regarding:
1. No guard
The construction was a learning project.  My choice of construction at least appears adequate for the pictured 16 fA current, besides a significantly higher pA current...
« Last Edit: March 04, 2017, 02:38:55 am by ocw »
Offline tggzzz
« Reply #80 on: March 04, 2017, 08:49:43 am »
I've picked up a couple of 1Tohm 3% "resistance transfer standards", without having anything to use them for, which implies TEA, FUSS, GAS, but since CAP didn't kick in, neither did DBA.
Anyway, on the side there is a typewritten sellotaped label
R1R2
100V1.0261.026
200V1.0271.023
500V1.0011.016
The corresponding 10Gohm 2% resistors lack such a label, unfortunately.
« Last Edit: March 04, 2017, 08:51:23 am by tggzzz »
Offline ocw
« Reply #81 on: March 05, 2017, 08:55:59 pm »
I'm trying to find the minimum current which I can measure with the 1T feedback resistor...
Besides the prior 1/10 and 1/100 voltage division, I added 1/1,000 and 1/10,000 options.  That allowed measurements of 1.6 fA and 160 aA via my ADA4530-1 current to voltage converter using its largest 1T feedback resistor.  The ADA4530-1_cal picture shows my converter with my grounded battery powered calibration box attached.  There's not much unshielded space for power line noise to influence the reading.  While the banana connectors aren't anything special, I'm using Teflon insulated wires.
The 1.6 fA measurement didn't look too bad after the reading stabilized--see the 1.6fA attachment.  Due to the large value resistors used with the small current, the stray capacitance in my circuit produced a significant time constant and delay before the reading became steady.  While the reading did stabilize close to the expected 1.6 mV output level for 1.6 fA, it wasn't steady enough for me to rate it as a success.  With the accuracy at 16 fA adequate to satisfy me, about 10 fA may be the limit for my circuit to have reasonable accuracy.  Based on my results in using increased value resistance, feedback accuracy and my recent results using the ADA4530-1 IC, it's likely if I paid $68.96 for a 10T resistor I would be able to have accurate readings down to 1 fA.
With the mentioned results at 1.6 fA, I wasn't expecting much luck in my final measurement at 160 aA.  I'm not used to dealing with attoamps.  As the 160aA attachment shows, it wasn't a complete failure.  The reading did center around 0 to the desired -0.16 mV to indicate 160 aA.  But, it only points to a very vague low current range.  The big jumps on both sets of today's readings could be due to power supply spikes.
I also wanted to take some readings using my ADA4530-1 circuit with my 1T ohm current limiting resistor along with up to 1,000 volts [with 2 mA current limiting] to take a comparative set of readings using my circuit's different feedback resistors.  The difference in the indirectly measured resistance of my external 1T resistor would partially depend on the voltage coefficient of the external resistor and also on the relative accuracy of my at least 1% accuracy feedback resistors (except for both 1T resistors being 10%).  I don't have any way of confirming which was more influential on my results other than what the 1% accuracy resistors suggests.  The results were:
  FB R  |   V to R    |   Voltage Out  |  Measured R
   1T       1.0000V        -1.0013V           0.9987T
 100G     10.000V        -1.0451V           0.9568T
  10G      100.00V        -1.0561V           0.9469T
   1G       1000.0V        -1.0915V           0.9162T
This has been an interesting project.
Offline PartialDischarge
« Reply #82 on: March 08, 2017, 07:24:19 pm »
The way I measure picoamps and very high resistances, great equipment  ^-^
Didnt found anything on this forum about this gear, so I will post a short review about it
Test resistor,
1 Volt across 100G:
« Last Edit: March 09, 2017, 03:30:02 am by MasterTech »
Offline ocw
« Reply #83 on: March 09, 2017, 09:19:18 pm »
While there are always good (and sometimes not so good) deals available, it seems like building a ADA4530-1 current to voltage converter has a hard to beat performance to price ratio for pA and below current measurements and G to T ohm measurements.  Using a much discussed U1252B meter to measure the converter's output, 10 aA (0.01 fA) current resolution is possible on the lower level currents.  One digit increased resolution is available with my 34401A.  Without zeroing either the converter or the meter connected to it, I had about a 2 fA current reading.  Zeroing the meter was only required for readings where that level of current was significant.
As the attachment shows I saw excellent accuracy with the resistors which I used as feedback resistors at 1gigaohm and below.  The 100 pA and above current measurement typically had around 0.3% accuracy and below.  My lower current accuracy verification was limited by the less accurate feedback and the external resistors which I used to set that current level.  I used a voltage level measured by a 34401A meter and a 0.01% 1M and below resistors and 1% resistors with values up to 100M.  I was limited by using a 10% 1T ohm resistor, which actually had a value close to 0.955T.
It was difficult to accurately verify the accuracy of even lower current measurements.
« Last Edit: March 10, 2017, 05:35:14 am by ocw »
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #84 on: March 20, 2017, 03:10:52 pm »
Let's see if people pay attention what I write. Anyone know where these beauties are from?  8)
Offline CalMachine
  • Metrology Nut
« Reply #85 on: March 20, 2017, 04:55:12 pm »
Your Keithley 6621!
All your volts are belong to me
The following users thanked this post: Vtile
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #86 on: March 20, 2017, 11:03:14 pm »
You are no fun.
Offline CalMachine
  • Metrology Nut
« Reply #87 on: March 21, 2017, 01:18:50 am »
Haha whoops!  ^-^   Very nice though!  How stable is that reading?
Offline ocw
« Reply #88 on: March 26, 2017, 01:41:48 am »
The attachment shows the output of my ADA4530-1 current to voltage converter when it is fed 250 pA AC current at 500 kHz.
At that frequency/current it has lost its conversion accuracy without taking its high frequency roll-off into account.
Offline ocw
« Reply #89 on: April 24, 2017, 03:42:30 am »
I recently completed the modifications of my ADA4530-1 based current to voltage converter.  A picture of it and my accuracy verification figures for it are attached.  It has the conversion ranges of from 1 pA input current producing 1 volt output, to 1 mA producing the 1 volt output.  Think of the current resolution which is provided in the 1 pA range.  Rather than the leakage resistance from my Johnson/Cinch banana input jacks or the Grayhill 56D30-01-1-AJS range switch, I had more problems at the lower currents due to noise and other currents coupling to any unshielded wires outside of its enclosure.  My alignment box solved things by having a N cell battery feeding a 1.0 volt regulator and a current limiting resistor inside of an earth grounded metal box.
My first version had two ranges with restricted frequency bandwidth.  Not finding them beneficial I added two non-decade ranges in my available switch positions.
The accuracy of the unit is primarily limited by the precision of the feedback resistors.  I have their values and accuracy shown on attachment.  The currents being measured are voltage dependent currents, NOT constant currents.  That's what I'm most frequently measuring.  I want to know what the worst case accuracy is, not what the accuracy is with ideal constant currents.  The current being measured was produced by a regulated 1.0000 volt and a current limiting resistor.  The accuracy of that resistor is shown on the attachment under the "Standard" "Accuracy" column.  So, the ADA4530-1 Accuracy column includes the insignificant amount of burden resistance/voltage in that IC.  The single digit fA input bias current of the op amp didn't effect zeroing the circuit's output until that amount of current became significant.  The external voltage meter can be zeroed if that value becomes important.
With an added external voltage or resistance and a simple calculation, resistance and voltage can also be measured.  The circuit has a very high input impedance for voltage measurement.  I've measured up to 1,000 volts using it.
I am surprised that a company doesn't use a similar circuit for their commercial circuit.  It would have a much smaller burden resistance and voltage for current measurements than what they are using now.  Greater value resistors could be measured.  And the circuit would have a higher impedance for voltage measurements.  I'd design the meter so that a 10M resistor could be added during voltage measurements for times when its high impedance causes more harm than good.  And, given the relatively low cost of the components needed, it could be used in a multi-meter with a cost significantly less than that of the typical electrometer.
« Last Edit: April 24, 2017, 06:34:00 am by ocw »
Offline Echo88
« Reply #90 on: November 13, 2018, 09:28:57 am »
Are there any recommended high insulation reed relays with >= 10 Teraohm, which are actually available and reasonable priced? Of course a few old COTO 1240 are available via ebay, but they are NOS. I was rather thinking about the HI-series from Meder, e.g. HI05-1A85, but those are non-stocked by Digikey/Mouser.
Maybe someone has measured a few common reed-relays which proved to have >= 10Teraohm and can make suggestions?
Reason for this question: Im building a multiplexed electrometer-box, to measure the leakage current of a few components automatically. At the moment im testing a few MEDER NP03-1A81-BV247, but they are barely reaching 1 Teraohm after a long time and id like to get better results.
http://www.farnell.com/datasheets/1486738.pdf COTO 1240 Datasheet
https://standexelectronics.com/wp-content/uploads/datasheet_reed_relay_HI.pdf
https://www.mikrocontroller.net/attachment/279340/3103181100D.pdf Meder NP03 Datasheet
Edit: Meder claims that the HI/CRR/CRF/KT-series are made to be high insulating, while the datasheets are very conservative with mostly "minimal 1 Teraohm spec": http://4a30d8fd18dae1bf393d-df49f4cedb726ad03ad145d2e3d346bd.r41.cf5.rackcdn.com/fileadmin/meder/pdf/en/Technical_Documents/Technical_Write-Ups/Electrometer_Detection_Circuits_Use_Reed_Relays.pdf
I will order a few of the CRR and KT-series-relays, measure them and then report back.
« Last Edit: November 13, 2018, 12:11:21 pm by Echo88 »
Online antintedo
« Reply #91 on: November 13, 2018, 02:43:37 pm »
I've been looking for them recently as well. HI05-1A66 is stocked by digikey. There are also a few others from the same manufacturer that have 10T minimum insulation.
Offline Echo88
« Reply #92 on: November 13, 2018, 03:41:17 pm »
Only the HI-series with the -85 suffix, like the HI05-1A85, claims 10 Teraohm insulation. And this type is not stocked by Digikey/Mouser. They are available from other US-sources, but cost ~20$ each. Therefore i will first test the CRR/KT-relays.
Is it possible to buy the reed-relay-casings (case with electromagnet and shielding, were you insert the reed-contact) somewhere? That would allow one to use good reed-contacts like the mentioned MAAR-5: https://www.littelfuse.com/~/media/electronics/datasheets/reed_switches/littelfuse_reed_switches_marr_5_datasheet.pdf.pdf
Offline e61_phil
« Reply #93 on: November 13, 2018, 04:27:13 pm »
I don't know if one can buy such housings, but I know a company, which provides 10T amplifiers which are switchable, very good ;). And they did it by their own, because you don't want to touch the sensitive net with anything.
Offline chuckb
« Reply #94 on: November 13, 2018, 07:15:59 pm »
I checked one of the MEDER reed relays (Digikey) for current leakage (PN HI05-1A66) with the K617 Electrometer. The DUT is suspended in air inside a metal box. Test conditions 23 deg C, 35% RH.
With 31V across the contacts I had 12 fa flowing. That worked out to about 2,600 Tohms. With 100V it was 42 fa. The coil to contact leakage was 4.6 fa with a 31v Bias. It looks like a very good high isolation relay.
The following users thanked this post: Echo88, antintedo
Offline serg-el
  • Country: ru
« Reply #95 on: November 13, 2018, 08:42:44 pm »
Offline e61_phil
« Reply #96 on: November 13, 2018, 08:45:29 pm »
Did you try to change the polarity to verify the measurement?
Offline chuckb
« Reply #97 on: November 13, 2018, 09:25:09 pm »
I did this test several months ago. I typically Suppress (zero) the 10 fa of offset current before applying the voltage. The zero is usually stable within 0.5 fa after the machine is warmed up and stable.
I will recheck the leakage across the relay contacts with both polarities later this afternoon.
The following users thanked this post: e61_phil
Offline chuckb
« Reply #98 on: November 13, 2018, 10:43:21 pm »
With +31 V bias the MEDER HI05-1A66 had 10.5 fa of leakage across the contacts. With -31 V bias it had -12.5 fa of leakage. So the average leakage at 31 V bias is 11.5 fa. This equates to over 2000 T ohms isolation resistance. I have previously measured the contact to coil leakage current being 1/3 of the contact to contact leakage current. This relay has been sitting on my bench for several month with no cleaning before this test. Hope this helps.
The following users thanked this post: Echo88, e61_phil
Offline chuckb
« Reply #99 on: November 16, 2018, 04:23:38 am »
I have also tested the  COTO 9012. This is a very small relay for less than $10 USD. The contact isolation resistance is 4000 T ohms. The coil to contact resistance is 1000 T ohms at 100V.