http://en.wikipedia.org/wiki/Kelvin%E2%80%93Varley_dividerA normal decade box is not a good idea. Your reference must see a high resistive constand low load. Your meter must see a constant resistance. .. A decade box will change impedance and the multimeter will form parralel resistance with the divider. The output of the KV will stay constant .
A KV is often used with a null-detector and standard cells. The KV was set at the value of the standard cell and the calibrator adjusted untill the null detector showed zero. In that case the calibrator was exact 1.1V, 10V or 100V and could be used for further tests. A Vref direct to a KV is possible if it is well buffered. I made a compound amp with a LT1052 chopper and LT1010 buffer (nummbers from the top of my head and I'm bad with that) it iis a voltage follower I use to buffer references before I connect some standards (I measure them with a 10 GOhm 7,5 digit to see if this is needed. My Fluke 731A surely needs it.
A decade is a waist of time and money for a reference.
If you use the last decade at 10, your 10 M multimeter is paralleled with 10 K. You can calculate the error. But now switch up to 9, so there is 90K parallel with 10M.
The KV will show the same resistance in every position.
Be carefull, the specs are often spread over a range. And the change you get one on the outerbounds is bigger as you think. They often sell several grades so carefull read the datasheets. Those letters often are important. The 1027 is available in A,B, C. I bought some C's at Farnell. I could not get a B. I then got a sample B from LT. I could not get the A version.
I have build a few standards. One with a LM399. That is now running 24/7 for a year. It drifted a lot at first. a few hundered uV. Then it jojo-ed plus/min 40uV a day (op 10V) but it does not drift any more. Changes are related to temperature and humidity. I monitored it a few weeks in the authum and noted the value together with rH% and tem in a graph. rH% seemed to give the biggest jumps. Tempco was good. It could jump 30 uV through tempchanges but that was over 5 or more degrees. so 6uV/C or 0.6ppm/C at 10V. But to my suprise, one of my first references made with 4 x TL431 parallel and a OP277 opamp performs very well. But not close to the LM399.
But beside that, your meter will react to to the same things. And if the meter is worst as the standard it will react more. If the meter is 10ppm/C in the 10V rangeit wil change 100uV every degree. So when the standard changes 30 uV over 5 C, the meter will change 500 uV. And if the accuracy is 0.001% that is 100 uV at 10V, so 10 ppm.
So if your reference is 0.6 ppm/C and centre excact 10.000,000V at 20C your meter must read between 9.999,900V and 10.000,100V. But at 25 degrees you should aid the tempco. and in our example +/500uV for the meter so the meter must read 9,999,400V and 10.000,600V. Then the 30uV from the reference. If you do not know the direction it will add that to the error budget.
These things can drive you (volt)nuts. I always warm the stuff up a day before I need it. Then compare everything to my standard cells to find out that if my roomtemp changes a few degrees I can start over again. So I must choose a day with constant temp, leave out my halogene lights (they heat up to much and cause EMC) and try to do a cal while watching hydro and thermometer.