I don't like STM, their marketing here in TW failed hard when I tried to get some support to consider using STM32 on a project.
Now let's get back on topic. I got this little cute PDVS2mini in thermal box, running
for life first quick and dirty tempco test.
I'm sure somebody will dig a photo with $70K vs $300 DC source
.
Design assembly is nice and tidy, with all critical stuff well cared. Graphical LCD is not backlit, just COG panel. It has decent contrast and main output voltage have nice large digits. The menu system is little unusual but not hard to figure out without any RTFM. There is a help string on the bottom strings to show keypad actions.
Access to battery requires removal of 4 screws, and also provide a way to see some internal construction.
There is no separate battery compartment or protection over electronic parts, due to low-cost targeting reason, which is understandable.
PDVS2mini designed to use two 9V batteries in series, with integrated charger. Unit needs to be powered on to allow charging from external DC jack.
I'm working on detailed review/testing, but it will take some time, so here are just some initial dirty results using my secondary equipment (K2002 and Arroyo TECpak controlled box).
PDVS2mini is powered by batteries, placed in DIY chamber. Raspberry Pi controls Keithley 2002 as a digitizer, and TECpak 585 for precision temperature control. 40W noname TEC used as cooler/heater, with water-cooling loop dissipating heat into room elsewhere. The sensor in the box is Honeywell HEL-705-1 100 Ohm platinum RTD. This thermal box setup may not look like much, but its able to maintain any programmed temperature from -5 °C to +65 °C with better than 0.008 °C stability over hours.
Now to the first test results. The first graph is a plot of +10V output from PDVS, with temperature change from +15.00 °C to +55 °C in a slow and controlled manner. There is soaking time at +35°C for 2 hours, and also at peak max temperature, to get an idea of settling and hysteresis behavior. The horizontal axis is time duration, vertical axis is deviation from initial point at +15 °C, which happened to be +9.999994 VDC +/-2ppm.
For illustrative purposes, chart have additional plots of -0.1 ppm/K and -0.2 ppm/K thresholds based on temperature change (using box method) and TEC current readout from Arroyo TECpak 585. Ramp up time for temperature is +0.08333 °C per minute. Measured deviation is -4.0 ppm for +15.0°C to +35°C ramp, and another -5.3 ppm for +35°C to +55°C ramp. Following the TC Box method we can estimate temperature stability at max
-0.23 ppm/K which is impressive for $300 USD DAC box!
Now same data, but in voltage/temperature scale representation.
Hysteresis on fixed temperature points is around ~2ppm but that will need additional verification with proper setup Unit may need additional time, longer than 30min to get temperature soaked to reach equilibrium. This particular temperature chamber does not use forced airflow, and from datalog looks like it need ~30 minutes to equalize temperature change.
Stay tuned for more
One part I'm not very happy about is banana-only output connectors. Once initial testing and calibration verification test is done, these will be for sure changed to some proper low-thermal EMF hardware
.
If somebody wants particular test setup/conditions - feel free to give feedback.