The quoted resolution on the 100 mV scale is 10 nV, but resolution only says so much. The accuracy is specified as 3 ppm of range, which is 300 nV. So leaving aside the offset for the moment*, the measured 30 nV difference isn't bad.
(* An offset at zero is virtually inevitable, and is typically subtracted mathematically, either internally (e.g. via a null button) or externally, though 3 uV feels a little high.)
Things are a little more difficult on the 10V scale. There, the accuracy is quoted as 0.05 ppm of scale, which is 0.5 uV. So 0.7-0.8 uV variation, which is under a factor of two away, isn't outrageously out of spec.
The LSB in the 100nV range is below 1nV for 10 / 100 NPLC at ~0.145nV @ 50Hz, although with the preamplifier noise etc., this resolution is of little to no use, so the 10nVs make more sense. The zero point itself is in spec, as +-1.06µV for 100mV / +-2.3µV are permissible for the 10V range (
@Zondar you might want to run ACAL to correct the general offset from zero, then you should also be in spec)
For the 10V range, what you call accuracy (0.05 ppm) is the transfer accuracy, so comparing two voltages in the same range. If the NPLC isn't changed during the measurement, this effect doesn’t matter as it affects both voltages the same way.
Where it will have an impact is with the absolute accuracy, which is specified with 0.5PPM / 24h in the 10V range, it will likely still be in spec, tho.
Although if we look at the additional gain error graph from the datasheet, we also have some breathing room here.
In conclusion, I would say that this issue wouldn't negatively impact our goal to hit the original 3458As specs.
Even with this problem, if my logic isn't wrong, a NU180-equipped machine might claim to be "8 digits", similar to where the slightly-elevated noise would put it. Does this make sense?
I think we could definitively give the project an 8 digits rating, as even with our current elevated noise levels, we are about as noisy as a Keithley 2002 and following wanghars linearity measurements, also the linearity looks quite decent and is likely in spec.
The only thing that is still problematic is the TC, which at this point is poor, even for a bad 6.5-digit meter, although wanghars experiments with foil resistors have also shown great results in this regard.
I will play around a bit more with this zero issue, but even if we don't find the solution, I wouldn't consider it to be a dealbreaker.
Greetings,
Simon