Author Topic: Sourcing nanoamps and below like a Samurai, or KEI 6221 DC/AC SMU teardown  (Read 11652 times)
Offline voltsandjolts
Re: Sourcing nanoamps and below like a Samurai, or KEI 6221 DC/AC SMU teardown
« Reply #25 on: January 19, 2024, 01:04:09 pm »
10aA source&measurement  :)
Approximately 62 electrons per second!  8)
Offline VintageNut
« Reply #26 on: January 25, 2024, 05:43:55 pm »
I am scratching my head for an application that requires the measurement of 10aA.
The noise of the Keithley 6430 is around 380 aA p-p. This was an actual measurement. The 6430 is arguably the best bench instrument in existence for sourcing and measuring low currents.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
Offline Kleinstein
« Reply #27 on: January 25, 2024, 06:11:27 pm »
There are not many measurements that need / care about such small currents. This a regime for speciallized instruments / labs. The Keithley 6430 may be best from Keithley and US general purpose sources. The meters shown are from Soviet time special scientific instruments build mainly for users like national metrology labs and similar. Alone the connectors with saphire isolators make them quite expensive (and also delicate) - a bit more than the teflon isolated triax found in more general purpose ones.
In this area the Soviets were quite good, sparing not much expense in handcrafted systems for special use. The Keithley meters are more the afforable version for every day use in the semiconductor industry.
One use for very low currents and resolving them accurate are electron turnstile systems - kind of trying to count electrons to get the ultimare current reference. This way one can provide small currents with good accuracy, but with the difficulty to transfer to more managable sizes.
Online bastl_r
« Reply #28 on: January 25, 2024, 06:17:13 pm »
Maybe Tunnelrastermikroskop.
Sorry, i don't know the english name.
Regards
The following users thanked this post: Samogon
Offline Kleinstein
« Reply #29 on: January 25, 2024, 07:06:57 pm »
A tunneling microscope would still use a significant higher current (still low, but more pA or even nA), as they need to be realtively fast to still get a picture with some 100000 and more pixels in a reasonable time. The very low current also implies a rather slow operation and long waiting times from parasitic capacitance and the associated dielectric absorbtion.
Offline VintageNut
« Reply #30 on: January 25, 2024, 08:36:08 pm »
If I understand correctly, the application is metrology. Specifically, measuring as close to zero as can happen with careful practice.
I have played with similar for resistance. Being able to measure in the 10 to 20 nano ohms region requires exactly the correct instruments and very careful practice.
I still have my ohms labs zero ohm standard. It is fun to try to measure.
Online Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #31 on: January 25, 2024, 11:12:02 pm »
Mass-spectrometry is one application which I am very familiar with. For example, these guys measure down to aA (the name of their patented amplifier is ATONA  = aA to nA range ),
Cheers
Alex
Offline VintageNut
« Reply #32 on: January 26, 2024, 12:13:52 am »
Cheers
Alex
Thanks. That helps. Looks like the instrumentation may be buried in the big iron.
https://www.isotopx.com/product/tims
Offline MegaVolt
  • Country: by
« Reply #33 on: January 30, 2024, 09:18:07 am »
How do they do it?
Offline Kleinstein
« Reply #34 on: January 30, 2024, 10:30:42 am »
I know of one masspectrometer front-end, that use a integrator at the input and photodiode as additional feedback. There the integrator is also directly part of an ADC.
Other very sensitive mass spectrometers use channe-plates as a kind of electron multiplier, a bit similar to a photo mulipliers. The output is than counting pulses, not really measuring the size of the current.
The following users thanked this post: MegaVolt