Electronics > Metrology
Precise and accurate measurement of high value resistors
ivo:
If you have two good meters, you can use voltage divider effect to measure the high resistor. Measure first meter's impedance in its volts mode, probably 10M nominal but of course you want to know that exactly. Then measure some very stable voltage exactly. Then you can put first meter underneath the 50M resistor in series connection with this voltage. With a 10V source you would get in the neighbourhood of 1.67V reading by the voltage divider formed. Work backwards to determine the exact 50M value. You can test with other smaller resistors to get a sense of the accuracy or determine viability of method.
Alex Nikitin:
Just in case if someone needs a set of reasonably accurate and stable resistors from 10K to 10G
Or you can just use the idea and build your own ;) , I use such a box for many years and find it very useful.
Cheers
Alex
MarkT:
Up in the 10G range I would worry about leakage currents - you need proper quality coax and connectors (polythene or PTFE insulation), and low humidity. Often a faraday cage is useful to keep out stray electric fields.
HighVoltage:
We had a similar discussion here:
https://www.eevblog.com/forum/metrology/my-triax-cable-shielded-box-project-for-keithley-instruments/
It is not that easy to measure high value resistors.
KK6IL:
Back in the 1970's, I was measuring insulation resistances up to 10^12 by putting the unknown R in series with a 1M 0.01% R and the pair put across an accurate 1000V from a Cohu 324A voltage calibrator. Leakage current created a voltage across the 1M resistor, which was measured with a Wavetek 704 0.01% differential voltage. A lot more accurate than I needed to test insulators but that's what I had to work with.
There's a limit on what one would be willing to spend to work on an old VOM, but an old Fluke differential, even if long out of cal would likely be accurate enough to implement the above test.
BTW, mounting a copper wire in an aluminum block does not create an insulated terminal but instead, a battery.
John KK6IL
Navigation
[0] Message Index
[*] Previous page