Would I be ok with this? No, but a few things must be considered. Here's some suggestions:
1) Same cables, Pomona or equivalent.
2) Set units to NDIG 8
3) ACAL ALL, wait 30 min to stabilize.
4) NPLC 100, or higher.
5) Look at calibration sheets. 4ppm/year is opt02, 8ppm/year drift is standard.
6) Two standard units can be up to 16ppm apart, without considering calibration uncertainties.
7) The units will drift closer and farther apart, several ppm over a day. Watch them and take readings every 30 min or so and see.
You need the absolute value of the reference, and the values from each unit's calibration to really assess the situation. On the face of it, yes, those 2 meters can be within their one year spec with those values.
PS - Most calibration houses will not adjust a meter that is in spec unless specifically requested to do so by the customer. If you want the numbers to match up, you have to request adjustment to 24 hour spec. When I send our master 3458a to Keysight, I ask for calibration to 24 hour spec only if the meter has drifted more than 2ppm or so above their reference values.
TomG.