Yes, i also found it confusing. Until i noticed both groups have their own 2nd key. When a 2nd function is active, the corresponding panels 2nd key has its indicator led on. In order to measure mADC the VDC and the left 2nd indicator should be on. There is only one range = 2000 mA. Input on right connectors.
The ratio function should really be in the left group. You terminate the ratio function by selecting another function, e.g. VDC. Similarly the instrument leaves the compute mode when pressing the VDC key on the left. When you are in AC mode and press null, it switches to VDC to null the instrument and then goes back to VAC.
What i am missing is a way to get the instrument into continuous measurement mode after going local. Could be some reset button. It's a pitty that product was abandonded.
With my research on the nonlinearity near zero issue our 6048 proved to be rather inert to mods, which is something good. Their conversion method is rocksolid, except i have my doubts about integrating digital control of the integrator (counters, pll control) together with analog control (amplifier and comparator). Also i have some doubts about running the ADC without negative supply. The only negative supply comes from U9 to pin 11. The comparator won't fire on the positive glitch of the integrator when starting run-down. There seems to be a minimum run-down duration of about 7 usec, in case the comparator triggers early. In that occasion summing/averaging errors happen. I found a capacitor missing in the schematic and another one present in the temperature ADC but optional and not present in the main ADC.
The nonlinearity effect amounts to about 90 to 200 pA error current at the integrator summing point. With 10 pC charge injection and 10 msec integrator cycle time, charge injection offset current can be 1 nA. Cutting into that at short pulse lengths can produce that nonlinearity. I am still waiting for the DG308B switches with 1 pC charge injection spec.
Regards, Dieter