Hi all,
Just looking to stir a bit of conversation over these resistors, which are available very affordably out there for the metrology turbo-hobbyist (the only kind).
I think they have accrued quite a bit of bad rap, from what I'm seeing based upon a critical post or two, and I'm just not sure I'm convinced this reputation is deserved.
I've recently acquired a set from a Tektronix lab that (apparently) closed shop relatively recently and they were measured within a few, very few, ppms of nominal in the easy values. They're also obviously well aged, and btw, aging served them very well if I'm going by the values measured by the lab. Just about uncannily well.
It's also what I'm seeing in my environment - they're within ppm of nominal (up to 20-25, depending on value) - again, almost uncannily so. By spec, tempco is 5ppm or less - again, very good.
I got another one since the original batch (to get a complete set of values) from a large US instrumentation retailer - for a song and a dance, even if carrying a verification sticker by the retailer from 2020 - and it also seems to be pretty much dead on. More nefarious impact from bad reputation, if you ask me. Not that I'd mind getting these for such little dough.
All this is measured with a calibrated 8502A, for anyone wondering. Checked with a "noncalibrated" (but exact, afaict) 8840A which confirms its readings.
Why the bad name for these? Is this what others have seen with these ESIs? I'd be very surprised, but maybe I just got very lucky with my set.