Is that M71 and HighVoltage's F732A? They differ by 1.2ppm?
The Fluke 732A (Wulf) was calibrated to 10V +0.6ppm in Stuttgart
The 7000 (M71) was calibrated to 10V +1.7ppm
10V + 0.6ppm + 12µV = 10V + 1.8ppm
The PTB 3458A agreed within 0.2ppm
I would say, that is a very good result