After understanding a bit better now how the instrument works out, the battery backed up memory is rather unimportant. There is no calibration data of meaning saved anywhere. And the machine do not get adjusted, just "aligned" which is some adjustments on some potentiometers to bring certain voltages on the boards to measurements with quite wide tolerances.
I still have not gone though that but will do.
Focus now is testing. Verified so far that I consistently can do 0.1 PPM or better 10K ohm transfers. First measure resistor A vs B, then B vs A and compare ratios. 0.15 PPM is the verification test limit, so a pass there. Amazing for something that has not been adjusted, calibrated or "aligned" in probably more than 15 years. I will do 10x ratios, high voltage, low ohm test etc later.
Its a rather interesting and exotic machine.