I have three related questions:
(a) What are your go-to switching matrices for switching with better than 10uV thermal EMF?
(b) Does anyone have any leads on the documented behaviour of relay contacts / contact materials when switching low voltages / currents?
(c) Failing off-the-shelf solutions, are there any go-to designs for low voltage switching (i.e. uV with max 5V) into high-ish impedances (i.e. 10M ohms)?
I'm looking for 25, 50 and 64 channel matrices - but only the 64 channel is critical. All leads welcome :)
TIA, Alan
“A foolish consistency is the hobgoblin of little minds"
Online Zucca
  • EE meid in Itali
Re: Relay / fet matrices for low current / voltage switching?
« Reply #1 on: August 14, 2018, 08:38:34 am »
What are your low current spec?
Can't love what you don't
« Reply #2 on: August 14, 2018, 09:09:10 am »
Good question :) and we need do to some tests - this is an opt101 photo-diode but with v. low light levels on particular IR wavelengths.
Alan
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #4 on: August 14, 2018, 11:00:58 am »
Thanks Alex.
FYI, for anyone else looking at this in the future, there's a doc here which gives a bit more analysis of relay contacts re low thermal:
https://www.mouser.com/pdfDocs/switching-low-offset-voltages-training.pdf
Alan
The following users thanked this post: Echo88, Zucca
Offline Kleinstein
« Reply #5 on: August 14, 2018, 11:54:46 am »
With so many channels an a rather low voltage, I would consider CMOS switches in IC form (like chips for a 8:1 MUX or even 16:1 MUX). The big down-side is possible leakage current.   Due to usually very low power consumption and a compact design temperature differences can be kept low.
Switching between photodiodes usually is not that critical with respect to voltage, it's more like a current sensitive case.
Offline Echo88
« Reply #6 on: August 14, 2018, 12:40:22 pm »
Thats the most comprehensive document on low-thermal-relay-construction ive ever seen and it even states the recommended manufacturer test procedure. Thanks.  :-+
    • A.N.T. Audio
« Reply #7 on: August 14, 2018, 02:27:29 pm »
Alan
From the datasheet that photo-diode + amplifier combination is rather noisy, with 300uV RMS noise in 20kHz BW (for a comparison the 1M feedback resistor thermal noise in the same bandwidth is ~18uV at room temperature). If you switch the amp outputs you will have more problems with the amps noise, offset and drift than with thermal voltages IMHO.
Cheers
Alex
Offline rhb
« Reply #8 on: August 14, 2018, 03:34:21 pm »
Has anyone documented the thermal EMF of the 44421A and 44422A relay boards from the 3497A?
Offline martinr33
« Reply #9 on: August 15, 2018, 12:55:38 am »
The quad CMOS switches that seem to populate precision meters might help:
https://www.maximintegrated.com/en/products/analog/analog-switches-multiplexers/MAX333A.html
I'd be more worried about thermals around the connections.
Keithley scanners /meters with scan options are reasonably inexpensive. Might be possible to switch out the relays with CMOS switches (maybe a little patch PCB). Then, you'll have a manageable package. 10uV isn't too hard to deal with.
There are not so many of them on eBay right now, but they do show up.
The bigger scanners have a lot more channels. The thermal offset is kind of in range as well.
https://www.conres.com/test-equipment/wp-content/uploads/sites/4/2014/04/selector_guide_7001_7002.pdf
Offline feedback.loop
« Reply #10 on: August 15, 2018, 04:09:54 am »
How about Agilent 3499A/B/C + some multiplexing module like 44470A?
3499B is a small unit with two slots for modules. 3499A is a bigger unit with five slots.
They can be found for under $100. Modules are also cheap enough.
Typical thermal EMF of 44470A relay module is specified as < 3uV
Be warned that many modules are sold on ebay with no terminal blocks.
« Reply #11 on: August 15, 2018, 04:14:25 pm »
As ever, some quality insight and experience :) many thanks to everyone who gave advice.
For speed, I think we'll try to pull in some off-the-shelf matrices and then develop our own custom ones as and when we've been able to characterise a bit better.
Probably 2 off 34972A plus 4 off 34908A 40-channel (3uV) or something similar.
Good point Alex about whether we're being too ambitious on the scanning given these particular photodiodes.
I'll keep you posted.
Alan
Offline splin
« Reply #12 on: August 16, 2018, 12:27:28 am »
That won't cut the mustard - the spec shows a typical ON leakage of 200pA; combined with the 10M source impedance requirement that gives a 2mV offset! And that will vary drastically with temperature.
The best such part I know of is the TI MUX36S08 with a typical ON leakage of 8pA. Still too high for this application but the leakage is a function of the input voltage relative to the supply voltages so actual performance may be better in this case. The scales of the leakage v temperature diagram in the datasheet are not helpful for room temperature applications but a more expanded graph is shown in figure 5 here:
http://e2e.ti.com/blogs_/archives/b/precisionhub/archive/2016/02/19/does-a-low-leakage-multiplexer-really-matter-in-a-high-impedance-plc-system
The 2N/PN/SST4117A Jfet (< 1pA leakage) might just do it but they are expensive, hard to get and require a lot of support circuitry. The MBM4117 is cheap but has a worse leakage spec - but possibly only because it isn't production tested.
Offline EmmanuelFaure
« Reply #14 on: August 16, 2018, 04:39:28 pm »
MAX326/327 : Max leakage 10pA, typical < 1pA.
Offline Kleinstein
« Reply #15 on: August 16, 2018, 08:44:04 pm »
The max326/327 are low leakage, but also relatively high resistance. It dependents on the application if this is ok.  The mux36S08 looks good - low leakage and much lower on resistance.