EEVblog Electronics Community Forum

Electronics => Metrology => Topic started by: Circlotron on December 11, 2024, 06:56:13 am

Title: Rigol DG1022Z frequency calibration guide
Post by: Circlotron on December 11, 2024, 06:56:13 am
I have a Rigol DG1022Z function generator I bought new 7 years ago and when I get access to a GPSDO I would like to tweak the frequency accuracy of the Rigol. Don't want to touch anything else, just the frequency. Could someone point me to a step by step guide how to do this?

I don't think it has noticeably drifted, just want to make sure it is as accurate as possible.
Title: Re: Rigol DG1022Z frequency calibration guide
Post by: Freddy15 on December 11, 2024, 09:54:51 am
Hi

Did that with my DG1062Z last week, see page 2-3 in the cal manual attached.
Suggest to use a scope instead of the frequency counter.
Plug your reference clock to one channel and the properly configured  CH1-out to the other channel.
Adjust so that both waves don't move against each other, will result in a mHz accuracy.

Cheers M.
Title: Re: Rigol DG1022Z frequency calibration guide
Post by: Circlotron on December 11, 2024, 01:11:20 pm
Thanks heaps Freddy15.  :-+
To enter the calibration mode it says to enter the password. Is there a default password for these?
Title: Re: Rigol DG1022Z frequency calibration guide
Post by: Freddy15 on December 11, 2024, 04:43:59 pm
Hi

If you haven't changed it  ;)  the standard calibration password is “123456” (see 1-1)

Cheers