I think you're right to keep the current points 5-5 centred and far apart
to give the current time to spread evenly. The right hand 5 as a point seems very close, I'm sure if you moved it to the top right or bottom right corner the reading would change, so keep it near the right hand center edge, I assume it will eventually clamped all along that edge which you might have to arrange for the maximum accuracy.
There will be no test current flowing and so no volt drop from point 2 and a half, to point 4, so that's where to measure.
When getting a 3 digit reading like 'abc', the last digit is always somewhat unreliable, so testing at 10A should give you an extra digit 'abcd' where 'd' is now the unreliable digit, but all abc are now accurate.
I don't think it's a very good shape either, looks like there will be too much of a concentration of current at the 2 bottom center corners to me.
There's nothing in the picture to show us the size or thickness of it, which would have helped with the milli/micro confusion, - I adjusted the contrast looking for a clear fingerprint!
Having said all that I think you're just about there.