Re: Teardown: Standard Resistors
« Reply #100 on: April 08, 2017, 08:59:27 pm »
I believe that up until the 80's standard resistors were supposed to be used with mercury wetted contacts. After all, they sold mercury cups/posts for that purpose.
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Teardown: Standard Resistors
« Reply #101 on: April 08, 2017, 09:08:22 pm »
Iih! :(

I was only afraid of cadmium solder until now...
« Last Edit: April 09, 2017, 11:55:33 am by e61_phil »
 

Offline Henrik_V

  • Regular Contributor
  • *
  • Posts: 155
  • Country: de
  • “ground” is a convenient fantasy
Re: Teardown: Standard Resistors
« Reply #102 on: April 09, 2017, 11:54:03 am »
When you hunt for the next digit in metrology, physics don't care human biology :-//
Cd in solder, Hg on contacts , my amatures are Be  ...   

Greetings from Germany
Henrik

The number you have dialed is imaginary, please turn your phone 90° and dial again!
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Teardown: Standard Resistors
« Reply #103 on: April 09, 2017, 12:28:39 pm »
When you hunt for the next digit in metrology, physics don't care human biology :-//
Cd in solder, Hg on contacts , my amatures are Be  ...

Pure Be oder BeCu?

If you don't inhale Be dust it is quite safe, isn't it?
 

Offline Henrik_V

  • Regular Contributor
  • *
  • Posts: 155
  • Country: de
  • “ground” is a convenient fantasy
Re: Teardown: Standard Resistors
« Reply #104 on: April 09, 2017, 04:40:50 pm »
I calibrate accelerometers and the amarture of the shaker is ceramic or mostly pure berillium. No problem unless you lick it for cleaning  ;)
Even our workshop got used to it, because after some years you need to replolish the armature, and you don't what to inhale the dust.  (Primary calibration with laser doppler interferrometers) And these workshop guys are awesome.  :-+ They also polish the Si-'orange' for the new representation of the kilogram. Different story...

From Wikipedia: https://en.wikipedia.org/wiki/Ohm
The "international" ohm was defined as a mercury column 106.3 cm long of mass 14.4521 grams and 0 °C at the International Electrical Conference 1893 in Chicago.
Greetings from Germany
Henrik

The number you have dialed is imaginary, please turn your phone 90° and dial again!
 
The following users thanked this post: 2N3055, CalMachine

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #105 on: May 05, 2017, 03:31:40 am »
Well... I'm back at it :-+  This time the magical RS925D!   :scared:

This was purchased, for pretty cheap, in an attempt to get an ESI 242D system up and going.  I opened her up to reapply a new foam seal as well as make sure there are no obvious issues that needed to be addressed.  After some quick preliminary tests, it looked like some of the lower decades had slight a hysteresis.   The issue seemed to have disappeared after letting the decade, and the K2002 I was using, thermally stabilize together for over 24 hours.





Let's open her up!


100 kOhm decade


10 kOhm decade


1 kOhm decade


100 Ohm decade


Top and Insides of trim pot assembly PCBs and WW resistors for the 100 kOhm, 10 kOhm, 1 kOhm, and 100 Ohm decades











10 Ohm decade


1 Ohm Decade




100 mOhm decade



10 mOhm decade.  These units have a fixed minimum resistance of 10 mOhms.  Because of this implementation there is no 'zero ohm', similar to insertion loss, measurements need to be taken into consideration.  This allows for the displayed resistance value to actually be the absolute value.



100 µOhm Decade fine trim pot


More Goodies!

100 kOhm and 10 kOhm decades




10 Ohm and 1 Ohm decades



1k and 10 kOhm decades
« Last Edit: May 05, 2017, 03:40:34 am by CalMachine »
All your volts are belong to me
 
The following users thanked this post: quarks, ManateeMafia, dr.diesel, doktor pyta, TheSteve, schmitt trigger

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 730
  • Country: us
Re: Teardown: Standard Resistors
« Reply #106 on: May 05, 2017, 03:41:25 am »
Thanks for sharing. It looks like it in great shape.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #107 on: May 05, 2017, 03:45:15 am »
Thanks for sharing. It looks like it in great shape.

You're are very welcome!  This is my first RS925D I've seen, so I wasn't 100% positive on the condition.  It seems as if there is a decent amount of flux residue still on some of the connections... is that normal? 
All your volts are belong to me
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Teardown: Standard Resistors
« Reply #108 on: May 05, 2017, 04:12:49 am »
Auch, purple wire resistor on 10K decade looks damaged to me. I have some Fluke PWWs, and those edges are real easy to get damaged and open wires, due to very fine gauge.
Thanks for sharing.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Teardown: Standard Resistors
« Reply #109 on: May 05, 2017, 07:42:10 am »
here are pictures from my RS925A for comparison to the RS925D above


« Last Edit: May 05, 2017, 08:27:17 am by quarks »
 
The following users thanked this post: CalMachine

Offline manganin

  • Regular Contributor
  • *
  • Posts: 189
  • Country: fi
Re: Teardown: Standard Resistors
« Reply #110 on: May 05, 2017, 09:31:52 am »
This time the magical RS925D!

That is a late one, seems to be from the mid 1980s.

Switches are the weak point, inferior compared with the General Radio counterparts. ESI solved the problem by paralleling several switch wafers. Which actually works fine, until dirty or totally worn out.

It would be a good idea to clean all the switches with a rough paper simulating the movement of the switch blade. You are ready when there is no dirt in the paper. Not easy to access and takes a lot of time, but worth the effort. Absolutely no switch cleaner chemicals!

Another thing worth checking are the screw connections of the slide wire. That is the 4-wire node so the possible change can only be seen in a 2-wire measurement. Open the screws, clean the surfaces and bend the spring washer teeth if required.

The trimmers are better that you would expect, but still the obvious reason for any high decade instability. That specific type is still available but very expensive. Turn back and forth a few times as a preventive measure.

 
The following users thanked this post: CalMachine

Offline Assafl

  • Frequent Contributor
  • **
  • Posts: 600
Re: Teardown: Standard Resistors
« Reply #111 on: May 05, 2017, 10:13:39 am »
Why not use a switch cleaner (like Caig stuff)?

For electrometers and the like one has to ensure that any cleaner does not touch any isolators such as Teflon, but if done extremely sparingly (e.g. Just on the wiping metal surface) then it does do away with oxidation and the like.

Does it affect calibration if it does not touch insulation?

(By sparingly I mean with a syringe or a wipe, and watch oilers do work well to dispense tiny drops into leaf contacts. )
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #112 on: May 05, 2017, 12:22:27 pm »
Auch, purple wire resistor on 10K decade looks damaged to me. I have some Fluke PWWs, and those edges are real easy to get damaged and open wires, due to very fine gauge.
Thanks for sharing.

I didn't really notice that at first, Good Eye!  Is the only coarse of action to repair said damage, to replace the resistor?


This time the magical RS925D!

That is a late one, seems to be from the mid 1980s.

Switches are the weak point, inferior compared with the General Radio counterparts. ESI solved the problem by paralleling several switch wafers. Which actually works fine, until dirty or totally worn out.

It would be a good idea to clean all the switches with a rough paper simulating the movement of the switch blade. You are ready when there is no dirt in the paper. Not easy to access and takes a lot of time, but worth the effort. Absolutely no switch cleaner chemicals!

Another thing worth checking are the screw connections of the slide wire. That is the 4-wire node so the possible change can only be seen in a 2-wire measurement. Open the screws, clean the surfaces and bend the spring washer teeth if required.

The trimmers are better that you would expect, but still the obvious reason for any high decade instability. That specific type is still available but very expensive. Turn back and forth a few times as a preventive measure.



Thank you for the advice!  :-+  I'll work on this here in the future.  Will probably post some before/after pictures, when I get to opening her back up.
All your volts are belong to me
 

Offline 0.01C

  • Regular Contributor
  • *
  • Posts: 91
  • Country: hk
Re: Teardown: Standard Resistors
« Reply #113 on: May 16, 2017, 05:38:44 pm »
I love 925D ,it is a great resistance stanard :-+
0.01℃
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 451
  • Country: us
  • The plural of anecdote is not data.
Re: Teardown: Standard Resistors
« Reply #114 on: May 17, 2017, 04:52:12 pm »
That discoloration on the 10K may be an indication of overload or a chemical spill, the coating on these resistors are not very resistant, even isopropyl will soften it up and if you get any on it, it will have to dry out thoroughly but has no lasting effects.  You'll have to check the resistor with a high accuracy measurement to see if it is out of tolerance.  You can use a contact cleaner like deoxit but do not spray it on the resistors, use a lint-less wipe.  Yes there is always some rosin residue left on the solder joints, it is of no consequence.  Usually these units require very little maintenance.

I have both a RS925D (part of my 242D) and a RS925A (part of a 242B) which I am going through and giving it a once over.  The connections on the rheostat are very important and should be snug (do not over tighten), the resistance element/wiper can be gently redressed with crocus cloth and cleaned.  I have found some units with some kind of gooey grease on the rheostat, that attracts dirt, I prefer to keep them clean and dry.  These units are more or less sealed in their case, there should be little to no contamination getting in there.

These resistors are wound on a mica card and therefor is somewhat easily damaged, not only the fine wire but the mica so be very careful around them.
 
The following users thanked this post: CalMachine

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #115 on: May 19, 2017, 06:54:09 pm »
She's Alive!!!!   :scared:

All your volts are belong to me
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Teardown: Standard Resistors
« Reply #116 on: May 19, 2017, 07:04:14 pm »
Is there a Helmut Singer Sticker?
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #117 on: May 19, 2017, 07:12:56 pm »
Is there a Helmut Singer Sticker?

Correct!
All your volts are belong to me
 

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2710
  • Country: us
Re: Teardown: Standard Resistors
« Reply #118 on: May 19, 2017, 08:10:51 pm »
I remember seeing one on ebay for $1500 obo, it sat on there for years. I think it was a dealer based on the West Coast.
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 730
  • Country: us
Re: Teardown: Standard Resistors
« Reply #119 on: May 19, 2017, 09:56:02 pm »
The black monolith is alive. Now the race is on to see which one of us can figure it out first. Mine keeps looking at me wondering when I will stop fixing stuff and start measuring my SR1010 and SR104 resistors.
I think there will be some projects to do soon involving resistance bridges.
 

Offline 0.01C

  • Regular Contributor
  • *
  • Posts: 91
  • Country: hk
Re: Teardown: Standard Resistors
« Reply #120 on: May 20, 2017, 12:37:26 am »
She's Alive!!!!   :scared:

How do you calibrate this black monolith? I have a 242C still sleeping on the floor. :palm:
0.01℃
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 730
  • Country: us
Re: Teardown: Standard Resistors
« Reply #121 on: May 20, 2017, 01:27:25 am »
According to the manual you need a 10k reference resistor with 1ppm calibration accuracy (SR-104) and a trimmable 10k/step transfer standard like the hard to find SR 1010 MTx.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #122 on: May 20, 2017, 01:34:19 am »
The black monolith is alive. Now the race is on to see which one of us can figure it out first. Mine keeps looking at me wondering when I will stop fixing stuff and start measuring my SR1010 and SR104 resistors.
I think there will be some projects to do soon involving resistance bridges.

You're on!   >:D I don't have any of the connectors or adapters so I ended up making my own.  We'll see how they turn out, as I have not put everything to the test yet.  I spent most of the day cleaning everything up on the outside, getting her looking good, as well as calibrating the 801.  The 801 performed very well and needed no adjustments coming from Helmut Singer. 

According to the manual you need a 10k reference resistor with 1ppm calibration accuracy (SR-104) and a trimmable 10k/step transfer standard like the hard to find SR 1010 MTx.

Done and done!   :popcorn:
All your volts are belong to me
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 730
  • Country: us
Re: Teardown: Standard Resistors
« Reply #123 on: May 20, 2017, 01:56:30 am »
Ok. Let's communicate offline and get some details worked out.

We can start a new thread and post any results if you want to do some detailed comparisons of portable standards.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Teardown: Standard Resistors
« Reply #124 on: May 20, 2017, 02:14:46 am »
Ok. Let's communicate offline and get some details worked out.

We can start a new thread and post any results if you want to do some detailed comparisons of portable standards.

Sounds good! And yes, I would like that very much. :-+