To all electronics freaks
After using the capacitance meter HP 4277A not exactly successful
to measure the self-capacitance of three electrometers,
I have checked the HP 4277A with a high-voltage air rotary capacitor with two independent capacity ranges.
Here is the high-voltage air rotary capacitor, turned in or
turned out:
The HP 4277 A allows measurement frequencies from 10 kHz to 1 MHz.
I therefore have measured both capacity ranges at 10 kHz and 1 MHz,
fully turned in or fully turned out, all values in pF.
10 kHz, large capacity range, fully turned in: 145.08
1 MHz, large capacity range, fully turned in: 145.49
10 kHz, large capacity range, fully turned: 43.30
1 MHz, large capacity range, fully turned out: 43.78
10 kHz, small capacity range, fully turned in: 54.38
1 MHz, small capacity range, fully turned in: 54.24
10 kHz, small capacity range, fully turned out: 22.68
1 MHz, small capacity range, fully turned out: 22.63
The HP 4277 A seems to work fine with the air condenser, the differences at 10 kHz and 1 MHz are very small.