Well perhaps.
Isn't everything that used the LTFLU, including the 8508A discontinued, why was LT still making it vs obsolescence? Or did Keithley find a pallet of these in Fluke's basement when the thought of the 7510 came up?
I presume LT was either still making them, or easy for them to do so, otherwise the LTZ would have been cheaper. Though I can't find a new price for the LTFLU anywhere.
Too bad for the secrecy, I guess we're not likely to find out more.
Why do you think the 8508A is discontinued? It is still available on Flukes site, no hint of obsolescence.
The LTFLU is also used in all actual Fluke calibrators and references, like 5730, 732B,
maybe very probably also in the 5790B, which are also not discontinued.
The LTFLU is a proprietary part, manufactured by Linear Technology exclusively for Fluke.
So no wonder you won't find a price, and can't buy it from them.
There's also no reason to believe that LT stops producing the LTFLU, like the LTZ1000, which is also still available, despite the limited sales volume.
Fluke for sure follows the policy to focus on the reference amplifier topology based on the SZA263/LTFLU type.
They also acquired Datron, including the 7001 reference (based on LTZ1000), and a few years later terminated this instrument.
That at first eliminated a competitor, but also saved cost, as their 732B was as good as the 7001, and they wouldn't have to maintain two different systems any more.
That includes all the validation / verification / monitoring for the references, where now they have to maintain one process and production line only.
Frank