An account of my attempt at in-house calibration of PRTs (platinum resistance thermometers).
This is the next logical step after my previously post about my reference resistor, where I concluded that my 6.5 digit DMM6500 has 4-wire resistance noise of +- 2 counts, corresponding to +- 0.5 mK for a PRT. The end goal is to characterize my homemade climate chamber and to calibrate sensors for atmospheric research.
The equipment is a DMM6500 with a 2000-SCAN card. This allows me to measure 5 PRTs in 4-wire resistance mode. My sensors are:
#1 PT100 Pentronic 10 cm probe, class B/3, probably wire-wound
#2 PT100 Labfacility 2420300 probe, class B, probably wire-wound
#3 PT1000 element IST P1K0.161.4W.B.010, thin-film
#4 PT100 thin-film Heraeus 32208550, Class A, Digikey 1759-1021-ND
#5 PT100 thin-film Heraeus 32208550, Class A, Digikey 1759-1021-ND
The sensor elements of #3-5 are merely 1-2mm in size, chosen for their fast response time in air. We cast their hair-thin legs and 4-wire solder points in epoxy with 3mm diameter for mechanical protection and electrical isolation against the water. The casting works quite well after some practice (except the cable stiffens from wicking up epoxy) but I don't know how the long-term measurement stability will be affected. Also, these are thin-film PRTs, instead of wire-wound which are said to be more stable over time.
Let's do an ice bath calibration!
This paper goes into great detail on precision ice bath measurements.
(NIST "Reproducibility of the temperature of the ice point in routine measurements"
https://www.govinfo.gov/content/pkg/GOVPUB-C13-5ce1be74727a73a2d8c89172d042b3a4/pdf/GOVPUB-C13-5ce1be74727a73a2d8c89172d042b3a4.pdf )
It shows the overwhelming importance of using ice from pure water to achieve 0.002 C accuracy of the melting point.
However, the calibration company Pentronic told me they use ice from (good quality) tap water, consistently getting +- 0.003 C. In my case, I used store-bought Norwegian ice and tap water.
As dewar, I used a glass vase ca 25 cm high and 10-15 cm diameter, wrapped in a bath towel. I used a consumer ice crusher to break the ice into pieces < 1 cm in size. A piece of styrofoam over the top helped to keep the sensors roughly in the middle of the vase. After inserting the sensors, I filled with ice, then with water pre-cooled in the fridge.
Playing around with the setup for an hour, the readings were mostly stable within 1-2 mK, with three caveats:
1. Some drifting occurred, especially in #3 which didn't seem to settle
2. Some sudden jumps between plateaus when stirring and siphoning off meltwater, notably #2 which lowered its plateau by 29 mK
3. Some individual readings are a fair bit (~ 20-30 mK) lower than the average, hinting that the average might be elevated
See attachments of multimeter plots for details.
For those interested, a cheat sheet for conversion between ohm and Kelvin for PT100:
1 K = 0.385 ohm
1 ohm ~= 2.597 K
My results are:
#1 Class B/3 (tolerance +- 0.1 K) Stable at 100.0116 ohm = 0.030 K above nominal
#2 Class B (tolerance +- 0.3 K) Stable at 100.0440 ohm = 0.114 K above nominal
#3 Class B (tolerance +- 0.3 K) Measured 999.23 ohm at lowest = 0.20 K below nominal. Too much drift to record full resolution.
#4 Class A (tolerance +- 0.15 K) Dips down to 99.9760 ohm = 0.062 K below nominal
#5 Class A (tolerance +- 0.15 K) Dips down to 99.9640 ohm = 0.093 K below nominal, but otherwise stable at 0.069 K below nominal.
All sensors are within tolerance. From the drift and dips in readings, my guess is that all sensors were exposed to the range 0 - 0.03 C at their lowest reading. This is a magnitude worse than the 0.003 C accuracy mentioned by the calibration company and I hope to improve on it. I made an earlier test with sloppier technique, getting less consistent results.
The process of stirring the ice and adjusting the sensors while looking for changing readings feels somewhat arbitrary (and time-consuming). Ice bath calibration is easy and cheap in theory, but my conclusion is that good results take some preparation and experience.
It will be interesting to repeat the test, attempt other temperature points and eventually send some sensors for 0.015 K calibration.
I'm learning as I go, comments are welcome.