Author Topic: Ultra Low Noise Reference 2DW232, 2DW233, 2DW23x  (Read 154727 times)
Offline 42Khz
Re: Ultra Low Noise Reference 2DW232, 2DW233, 2DW23x
« Reply #625 on: June 01, 2019, 02:42:18 pm »
Recently i bought 10 pcs 2DW232 from aliexpress.
Can anyone suggest me a simple method to test if them are originals or fake?
I have only 2 Multimeter (UT61E + AN850B), 1 Power Suply KA3005D and a Rigol DS2072A.
The easiest way is to crack one open.
If it's a single die inside covered in some gunk it's probably real (I'm not aware of any fakes like that) and if it's two discrete zeners it's a fake.
Offline Kleinstein
« Reply #626 on: June 01, 2019, 03:59:02 pm »
Looking inside is probably the easiest way.
A non destructive way would be to build a simple reference circuit and measure the noise. Chances are high the fake ones would be much higher noise. I would not mind "fakes" that are very low noise.
Offline stijena1973
  • Country: hr
« Reply #627 on: September 02, 2019, 12:13:29 pm »
I suspect that the component is fake. BE junction of some si transistors in the opposite direction behaves as 5.6 V zener. Could it be a mislabeled transistor?
Check for hfe, just in case.
Online sorin
  • Country: al
« Reply #628 on: September 02, 2019, 01:04:38 pm »
They are not  transistors.
Offline cape zoloh
« Reply #629 on: September 20, 2020, 09:05:25 pm »
Sorry to bump this thread. If anyone got some of the 2DW233/2 (with the diamond mark) for sale (or trade), please send me a message!
« Last Edit: September 20, 2020, 09:11:47 pm by cape zoloh »
Offline MegaVolt
  • Country: by
« Reply #630 on: September 21, 2020, 12:44:58 pm »
I bought here. There is a diamond sign.
Didn't check the parameters :(
https://aliexpress.ru/item/32871103478.html
Offline Pipelie
  • Country: cn
« Reply #631 on: September 22, 2020, 06:58:51 am »
be ware that not all the 2DW232 with diamond mark is the low noise one. we have tested samples from several source and we found the date code start with 12-xx and 13-xx are most likely the low noise one. the newer one is probably not, such as the 14-xx to 15-xx we tested are kind of bad.
The following users thanked this post: BravoV, edavid
Offline KK6IL
« Reply #632 on: October 17, 2020, 04:58:18 am »
I ordered 20 from the address MegaVolt supplied, and received 2 packages of 20 each, date code 15-4 (and 2 credit card charges). I may or may not have double ordered, not being used to AliExpress. Inked markings not very distinct.
With Pipelie's warning, I wanted a quick noise test so set up a comparison with a LM329. I set up a test using my LM399 standard as the source, 3K resistor for the LM329, and 750 ohms for the 2DW323, pins 1 & 2. Measured with Tek 7A22 plug-in set to 1Hz-1kHz bandwidth. Photos captured with Hantek DSO connected to 7834 Vertical Output. LM322 voltage source measures 35.6uV P-P. LM329 measures 25.2uV P-P, and 2DW232 measures 5.2uV P-P.
Offline Andreas
« Reply #633 on: October 17, 2020, 02:15:46 pm »
Hello,
so you have 1:50000 amplification?
Hmm, to measure 1Hz you would need a minimum sweep time of 1 second.
Otherwise you get another pole by the limited measurement duration.
with best regards
Andreas
The following users thanked this post: KK6IL
Offline KK6IL
« Reply #634 on: October 17, 2020, 11:43:39 pm »
You're correct. The high pass filter on the 7A22 plug-in ( love those) was set to 1 Hz, but the sweep time would have to be set slow to actually show really low freq noise.
I was just glad to see that the 2DW232's from that site were real Diamond components and were low noise compared to the LM329's.
I built a copy of a "low noise" 60 db amp, and was trying to add that to the test setup to get further above the scope noise when my LM399 supply started latching into a 3.5V output mode,  |O so will need to build a low noise, higher current supply to do further testing. The DW will be good for that as I will only worry about short term stability, and not 0 TC.
Offline Kleinstein
« Reply #635 on: October 18, 2020, 09:28:31 am »
Chances are that even the not so good 2DW232 would be OK at the higher frequency. The real test is for the low frequency range and 1/f noise. The 1/f noise is what depends one purity and can vary between batches.
For a simple noise test one would ideally have to independent references and than measure the difference.
There is no real need to have a separate low noise supply. The normal 1 OP amplification of the ref. and getting the current from the amplified voltage should be Ok. This may even work with an LM358 on a breadboard - though poor contacts can be a hassle.
The difficulty may be more in setting the right current as 1/f noise and temperature fluctuations can look similar.
Offline edavid
« Reply #636 on: October 19, 2020, 03:46:51 pm »
Pipelie, what is your noise test method?
Offline Pipelie
  • Country: cn
« Reply #637 on: Yesterday at 06:58:58 am »
Hi edavid,
the LNA I'm using is this one. https://www.eevblog.com/forum/metrology/diy-low-frenquency-noise-meter/50/
also check post #195 for user manual. https://www.eevblog.com/forum/metrology/diy-low-frenquency-noise-meter/175/
test method:
[attachimg=1]
result of 2DW234:
[attachimg=2]
and others