The likely most critical parameter is the long term drift and here data are very sparse.
So datasheets don't give us the information we would need for this purpose. Moreover I see not all OEMs use the same standard, and it seems some OEMs use more than just one standard.
I guess load life is the best we can look at? And, perhaps, moisture resistance?
Anyway i compared specs betwenn UPF, MMRF and PTF. Here it is FWIW.
UPF25 MMRF2568 PTF56
Tolerance 0.1% 0.1% 0.1%
TCR 5 ppm 5 ppm 5 ppm
Endurance/Load life 0.2%+0.05R 0.05%+0.05R 0.04%
Damp heat with load/Moisture resistance 0.2%+0.05R 0.05%+0.05R 0.08%*
Pulse/short time overload 0.1%+0.01R 0,02% 0.01%*
Resistance to soldering heat 0.05%+0.01R 0.05%+0.05R 0.02%*
Note: * = Typical Vishay