Thermocouple accuracy
After reading that Datron ageing paper ( thanks CaptYellowShirt ) Ive embarked on a oven to thermally work my refs & resistors for use ( and final ref boards ).
...
Lucas
Thanks to Brandon, I also read this interesting paper.
Anyhow, I still disagree to try to pre-age the LTZ1000, or hermetically sealed, oil filled metal foil resistors, if you use them for the reference.
The paper describes the fast ageing of 1N829 zener diodes, which are used in the Datron 4000, but definitely NOT the LTZ1000, which sits in the Datron 4910!
The big difference is, that the LTZ1000 already has such a small typical drift (1ppm/yr.) from the beginning, that further ageing makes no sense.
The 1N829 have typically 20 ppm/yr, and may be stabilized by that method to a drift level where the LTZ begins.
The LTZ on the other hand shows strong hysteresis of several ppm, if its chip is heated to high temperatures as 110°C, which may lead to a strong creeping drift behaviour in the following years.
Therefore, a heat shock leads to a less stable reference.
Same goes for metal foil resistors, the oil filled types have such low annual drifts, (and also hysteresis) that any heating will have negative influence on stability.
Frank