Hi Dick,
As far as commonly known, HP used the A version only, due to the nominal 95•C.
12.1k sets the oven to about 45•C but the interior temp already will be 35.. 40•C at least, so the difference is much too small, as the A version requires 10•C overhead plus some additional regulation margin, at least.
The self heating of the 3458A adds 13•C to room temperature ,plus 5 •C more for a dirty filter, plus 10•C if placed in a rack.
Therefore, I always recommend to set the temp to 65•C , ie 13k, or 100k in parallel to the 15k, if the lab is not warmer than 30•C
The reference ages only, when it is powered.
Therefore, drifts < 1ppm/year can be expected on 65•C, if the 3458A is not running continuosly.
If you use the ref board outside the 3458A, then 12.5k is the correct value, for about 55•C.
45•C is a little bit critical for the A version, otherwise you have to make sure, that the ambient temperature will not exceed 30•C. And there should be a thermal shielding around the TO 99, which should allow heat exchange, ie not too good an isolating material.
Frank