Author Topic: Using gold plated brass binding posts on a resistance reference box  (Read 10113 times)

0 Members and 1 Guest are viewing this topic.

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
I sometimes do multimeter reviews on the youtube channel and would like to have a resistance reference box to use when assessing the accuracy of these meters. They are usually cheap, low resolution meters (up to 4.5 digits) so I don't think I'm in the metrology type measurements area on the same principle where RF guys consider sub Ghz signals DC  :-// but still I thought this is the best place to ask. I also have two bench meters Agilent 34401A (6.5 digits) and HP 3478A(5.5 digits) and it would be great if I was able to keep these in check as well with this reference box.

Vishay sent me 3 resistors from their VHP100 series as free samples, they are 1K, 10K and 100K. The plan is to use an aluminium enclosure, drill holes for the binding posts (4-wire arrangement) and connect the terminals using some copper silver wire, by soldering on the ends of the binding posts.

Then I would have the VHP100 resistors, suspended and soldered in between the solid silver copper wires. I would be soldering them at least 10mm away from the can and I would also try to heatsink them while soldering to minimize the thermal shock applied to the resistors.

I've looked for good quality binding posts and I know the Pomona 3770 (Gold plated Tellurium Copper with polycarbonate insulation) is often referred to on this metrology sub-forum but unfortunately they are too expensive at $12.5 USD + shipping + tax, considering I would be needing 8pcs total. So instead I would like to use the Hirschmann 935 980-811, these are gold plated brass connectors and I got a set of 10 for about $32. 

I've read several threads on this forum, talking about the issue of EMF (Thermoelectric voltage) which appear at the junction of dissimilar materials however I would like to know how would this issue affect me with my particular build. What kind of errors am I expecting to introduce into my measurement if I use those Hirschmann gold plated brass binding posts in a 4-wire measurement setup?

Offline F64098

  • Regular Contributor
  • *
  • Posts: 53
  • Country: de
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #1 on: September 16, 2018, 12:12:05 pm »
Hello,

the BIL 20 AU have a very "cheap" look.
Why don't you look at PKI 10A AU?

Regards

Frank

 

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #2 on: September 16, 2018, 12:18:34 pm »
the BIL 20 AU have a very "cheap" look.
Why don't you look at PKI 10A AU?
I'm not sure if these have any technical advantage or it's just looks but I can say that I prefer the "slim" type look of the 935 980-811 and the fact that it has that split in the end for soldering connection.
Looking at the two datasheets I spot a difference on the housing material 935 980-811 quotes PF While the PKI 10A AU quotes PA. But once again I don't know if there is any advantage or disadvantage to any of those types of plastic.

Offline nanofrog

  • Super Contributor
  • ***
  • Posts: 5446
  • Country: us
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #3 on: September 16, 2018, 02:16:00 pm »
Given your stated usage, the BIL 20 AU would do just fine (Hirschmann makes decent stuff).  :)
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14073
  • Country: de
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #4 on: September 16, 2018, 04:42:56 pm »
For normal use the cheaper sockets are Ok. The main point is to avoid a temperature difference between the voltage sensing terminals. If mounted in an more insulating case, the sockets will follow the temperature of the plugs and thus not that much temperature gradient over the brass part anyway.  It also depends on the cables used.

For directly connection of just a 4 wire cable one could consider additional srew terminals wired in parallel.
 

Online splin

  • Frequent Contributor
  • **
  • Posts: 999
  • Country: gb
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #5 on: September 17, 2018, 05:01:31 am »
You only need low thermal EMF for the voltage sense terminals. Assuming you have a common terminal for the three resistors you only need four low thermal EMF connectors. The current force termnals can be the cheapest and nastiest that look ok.

With 4-W measurements I can't think of any reason why you would need silver plated copprr wires - just about any, cheap, wire will be perfectly ok.
 

Online tggzzz

  • Super Contributor
  • ***
  • Posts: 19280
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #6 on: September 17, 2018, 08:07:30 am »
Guildline 9330 resistors have copper terminals: https://xdevs.com/doc/Guildline/9330/Guildline9330Datasheet.pdf

My Sullivan T1212 has copper and mercury bath terminals.
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #7 on: September 17, 2018, 08:10:14 am »
With 4-W measurements I can't think of any reason why you would need silver plated copprr wires - just about any, cheap, wire will be perfectly ok.
I want to use the silver plated copper wire because it will not oxidize like copper would do after a while.


Offline 001

  • Super Contributor
  • ***
  • Posts: 1170
  • Country: aq
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #8 on: September 17, 2018, 01:27:41 pm »
hi

iz it real gold or some titanium oxyde cover?
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14073
  • Country: de
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #9 on: September 17, 2018, 04:29:45 pm »
I have not seen titanium nitride used for contacts. It looks golden and is conductive, but can be covered with a thin oxide that makes poor contacts. Resistance is also quite high so it is more like candidate for thin film resistors and AFAIK even used in a few cases.
A very thin gold layer is not that expensive. One sometimes even gets this with cheap transistors or similar.

Silver plated copper wire is not that exotic and it solders really nice.
 

Offline mmagin

  • Frequent Contributor
  • **
  • Posts: 610
  • Country: us
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #10 on: September 17, 2018, 11:14:22 pm »
I think the big thing to watch out for with binding posts not designed for low-thermal-emf is that under the relatively thin layer of gold is a thick layer of nickel, and nickel-copper makes quite a thermocouple :/
 

Offline 001

  • Super Contributor
  • ***
  • Posts: 1170
  • Country: aq
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #11 on: September 18, 2018, 05:11:32 am »
I have not seen titanium nitride used for contacts.

China banana terminals for example
It looks shiny gold but cheap
What about this "gold"?
 

Offline MiDi

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: ua
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #12 on: September 24, 2018, 12:32:51 pm »
voltlog, you are on the Foil news @ VPG  :-DD
 

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #13 on: September 24, 2018, 02:46:51 pm »
voltlog, you are on the Foil news @ VPG  :-DD
Where? I cant find myself.. I'm lost  :palm:

Offline MiDi

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: ua
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #14 on: September 24, 2018, 03:21:40 pm »
 
The following users thanked this post: Barryg41

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #15 on: September 24, 2018, 05:09:48 pm »
Oh, OK. That's their twitter feed. Cool! Thanks for sharing this.

Offline meandeev

  • Regular Contributor
  • *
  • Posts: 66
  • Country: de
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #16 on: September 26, 2018, 07:15:03 pm »
China banana terminals for example
It looks shiny gold but cheap
What about this "gold"?

And more: it´s not copper under the gold(?) - it looks like aluminium and you cannot solder it.
 

Offline MiDi

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: ua
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #17 on: September 27, 2018, 06:05:39 am »
the BIL 20 AU have a very "cheap" look.
Why don't you look at PKI 10A AU?

I thought about using the PKI 10 AU for my resistorbox, but they do not have soldering point and only give 2mm for case thickness.
So no good option for me as my chosen plastic case and metalbox (HF type) have around 3mm together.

For now I put an eye on the Keystone 4109 pair (brass, gold plated) for ~10$.
Btw what the hell is the difference between 4108 (20$) and 4109?
« Last Edit: September 27, 2018, 06:26:55 am by MiDi »
 

Offline czgut

  • Contributor
  • Posts: 15
  • Country: pl
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #18 on: December 21, 2018, 04:00:12 am »
Practical view:
EMF due to improper match of plug to jack metal is usually of a few  μV. It is important on 0.1V and below.
Your 34401A is using 1mA to measure 1kOhm, 100 μA to measure 10kOhms and 10 μA to measure 100kOhm.
All this cases give 1V Voltage drop on measured resistor. That means You will have temporary few microvolts thermal EMF (few minutes, until temperature between plug and jack will equalize). This introduces a few ppm (part per million) temporary error, visible on last digit. This is comparable to floor noise and instability of 34401A (5ppm of the range, according to specification). And few times less than 24h 20ppm stability.
So, if You do not like to wait few minutes until readings will stabilise, go into gold plated. Gold plated brass will be also OK.
If You want to use also Your resistors as Low Voltage divider, go to gold over copper tellurium. 
8.5 digit multimeters usually have TrueOhms function (reducing thermal (and any other constant) EMF by current on/off or reversal).
 
The following users thanked this post: TiN, voltlog, kj7e, wolfy007

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #19 on: December 21, 2018, 06:54:03 am »
Thank you for the practical view, exactly what I wanted to hear and also I learned a few things from this post  :-+

Offline czgut

  • Contributor
  • Posts: 15
  • Country: pl
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #20 on: December 22, 2018, 01:25:46 am »
Re nuts of choosing binding posts...
I would choose well made mechanically: with flat parallel surfaces, allowing for large contact area to spade lugs. Bigger contact area  ==> smaller and more repeatable contact resistance, + faster temperature equalizing between spade lug and binding post. Good binding posts should allow keep contact resistance below 0.3 mOhm and repeatability below 0.1mOhm. This is not important in case od 4-wire Ohms measurements, but why not to have good Reference Voltage Divider for the same money..
I would choose Metal case. It can be used as heatsink to resistors, and will equalize differences of temperatures between resistors and between binding posts.
 

Offline voltlogTopic starter

  • Regular Contributor
  • *
  • Posts: 101
  • Country: ro
    • VoltLog
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #21 on: December 22, 2018, 02:42:35 pm »
I actually used the binding posts which I had, mentioned in my first post above. I used an aluminium enclosure but I did not connect the cans to the outer casing.
Here is the build video if you would like to take a look at how I did it.

 
The following users thanked this post: Inverted18650

Offline RandallMcRee

  • Frequent Contributor
  • **
  • Posts: 540
  • Country: us
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #22 on: December 22, 2018, 06:10:32 pm »
Hmmm.  You should always heatsink those resistors when soldering them. 

Your nice box should have verbiage giving the tolerance as well [emoji41]


Sent from my iPhone using Tapatalk
 
The following users thanked this post: Inverted18650

Offline czgut

  • Contributor
  • Posts: 15
  • Country: pl
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #23 on: December 23, 2018, 03:08:51 am »
Nice work.
In case of using as Resistor Standards for Your Multimeter You do not have to add heatsink. But some Multimeters use 1mA to measure 10kOhms. Also, if You will use Your set as 1:11 voltage divider or as half of the bridge, and supply it  from 10V You may notice influence of selfheating. This power 10V x 1mA => 0.01W may heat Your resistor 1..2deg C, and change resistance 0.3  ..1ppm (depending on TCR of Your resistor).
For Your multimeter it is OK. But if You want to be ready for better multimeter You can add Cu or Al foil to make thermal shortage between resistor can and case.
 

Offline Inverted18650

  • Frequent Contributor
  • **
  • Posts: 875
  • Country: us
    • Test Gear Addicts
Re: Using gold plated brass binding posts on a resistance reference box
« Reply #24 on: December 23, 2018, 04:45:57 am »
Interesting build. Instead of paying big money for individual binding posts, you could find an old school 6.5 digit meter on ebay for less than $100 and scavenge them. An old Fluke 8502 has 7 posts and you can get them for less than $80 broken. Just a thought. Maybe go a step further and add oil to the inside of your case to ensure long term stability (though that case is rather large).  Stoked to see the future results of this project, thanks mate!
 
The following users thanked this post: voltlog