Author Topic: Very strange behavior of my Keithley 182 nanovoltmeter  (Read 3822 times)

0 Members and 1 Guest are viewing this topic.

Offline gamalotTopic starter

  • Super Contributor
  • ***
  • Posts: 1306
  • Country: au
  • Correct my English
    • Youtube
Very strange behavior of my Keithley 182 nanovoltmeter
« on: November 25, 2018, 08:02:45 pm »
I got a Keithley 182 nanovoltmeter with a display failure a few days ago, which is easy to fix by making a simple 60V boost converter. Then I did some simple tests on this instrument. Unexpectedly, I got a very interesting (strange) result.

In short, in the 30V range, near 3.25V 5.2V, the voltage measurement has a sudden jump about 100uV. I have repeated this many times in the past, and the results are the same, no matter how I change the filter, integration time, or other settings.

 :-//
« Last Edit: November 26, 2018, 11:51:54 am by gamalot »
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #1 on: November 26, 2018, 12:50:58 am »
Hi,

a couple of month ago I did some linearity tests with my K182. I attached the results. I can do further measurements if that helps (not before december).

First row is calibrator output voltage, second row is K182 reading and third row is temperature.

best regards
Philipp

Edit: The K182 isn't as bad as is looks here. The comparison is not fair in 1/3 of the range. Therefore I did another test in the 3V range with a divider. That showed +/- 0.4ppm. But that is offtopic, I think.
« Last Edit: November 26, 2018, 12:55:13 am by e61_phil »
 
The following users thanked this post: gamalot

Offline gamalotTopic starter

  • Super Contributor
  • ***
  • Posts: 1306
  • Country: au
  • Correct my English
    • Youtube
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #2 on: November 26, 2018, 11:51:18 am »
Hi,

a couple of month ago I did some linearity tests with my K182. I attached the results. I can do further measurements if that helps (not before december).

First row is calibrator output voltage, second row is K182 reading and third row is temperature.

best regards
Philipp

Edit: The K182 isn't as bad as is looks here. The comparison is not fair in 1/3 of the range. Therefore I did another test in the 3V range with a divider. That showed +/- 0.4ppm. But that is offtopic, I think.

Thank you Philipp!

Thank you very much for the reply, it made me find out a mistake in my post: it should be 5.2V not 3.25V. At the same time, your reply also proves that my K182's strange behavior is not a special case.

I will do some further testing to try to figure out what is happening near to 5.2V.

If you don't mind, can I know the firmware version of your K182?

 ;D
« Last Edit: November 26, 2018, 11:52:56 am by gamalot »
 

Offline gamalotTopic starter

  • Super Contributor
  • ***
  • Posts: 1306
  • Country: au
  • Correct my English
    • Youtube
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #3 on: November 26, 2018, 01:48:05 pm »
I did a sweep up and then sweep down, I was surprised to find that there is a hysteresis, just feels like auto ranging.


Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #4 on: November 27, 2018, 04:59:53 pm »
I will try this sweep next week

My K182 shows the following on startup:

Keithley 182 SDV
A04 B03 A01
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #5 on: December 13, 2018, 05:20:12 am »
I tested my 182-M last night too. It is modified version with 9V battery powered front-end. FW version A04 B03 A01. Calibrated earlier this year using procedure from manual and 5720A/Keithley 262 setup.
INL data I got leaves me wondering too:

-3/+3V sweep on Keithley 182M 3V range. 100ms integration, Filters off, 6.5 digit mode. Reference meters in 10V range



Correct data below:

« Last Edit: December 14, 2018, 11:21:40 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: gamalot

Online Vgkid

  • Super Contributor
  • ***
  • Posts: 2710
  • Country: us
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #6 on: December 13, 2018, 12:18:24 pm »
That is kind of weird.   ???
Thanks for posting. :)
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #7 on: December 13, 2018, 12:45:27 pm »
del some nonsense..

Doing test with 30V sweep currently. (other meters on 100V range).
« Last Edit: December 14, 2018, 11:33:04 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16615
  • Country: us
  • DavidH
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #8 on: December 13, 2018, 03:03:56 pm »
What type of ADC does it use?  That looks like the glitch commonly found in synchronous voltage-to-frequency ADCs.
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14196
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #9 on: December 13, 2018, 05:10:08 pm »
It's linearly nonlinear  8) . Perhaps I'll change the battery in front end and test again some time later.  :-BROKE

Doing test with 30V sweep currently. (other meters on 100V range).

The linear part (like in the 3 V range of the 182) of the "INL" curve kind of indicates a problem in calculating the straight line to subtract. So this is not a property of the meter, but a kind of test problem.  So the 3 V and 0.3 V range look good - with possibly some minor problems in the 2.2 V range.
 
The following users thanked this post: TiN

Offline gamalotTopic starter

  • Super Contributor
  • ***
  • Posts: 1306
  • Country: au
  • Correct my English
    • Youtube
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #10 on: December 13, 2018, 05:29:55 pm »
Is it possible that the input protection circuit causes this strange result?

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14196
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #11 on: December 13, 2018, 07:09:50 pm »
I don't see a problem with the input protection. There may be an issue with the extra 30 V amplifier, if it's only used for more than 5 V instead of for the whole 30 V range.
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #12 on: December 14, 2018, 11:31:41 am »
Kleinstein right on money. Sure, there was a bug in the data plotter. Shame on me for silly mistake. Array with INL points was calculated by using poly from channel 5 (F8508 data) instead of channel 6 (K182M).

Pulled wrong pics from post above, will replot the proper ones tonight.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #13 on: December 16, 2018, 12:40:04 pm »
Sorry, I forgot my promise to take some measurement.

At first here are some old measurements for K182 INL. Measured against my Fluke 5440B. The 5440B was sweeping from 0 to 10V for best linearity. The 0 to 10V was then reduced to 3 to 10 by a SR1010-1k.

@TiN: Could you reproduce the hysteresis?

Now I will start some up and down sweeps to see if my K182 also has this hysteresis.
 

Offline gamalotTopic starter

  • Super Contributor
  • ***
  • Posts: 1306
  • Country: au
  • Correct my English
    • Youtube
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #14 on: December 16, 2018, 01:09:27 pm »
Thank you guys! I am doing something really boring!   :=\

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #15 on: December 16, 2018, 07:20:24 pm »
I could confirm these hysteresis..
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #16 on: December 17, 2018, 04:03:01 am »
I see the jumps on 30V range too, at about +/-5.2 and +3.5 VDC levels.  :-//



On lower ranges like 3V and 300mV there are no problems (at least in my unit).  :-+





YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Very strange behavior of my Keithley 182 nanovoltmeter
« Reply #17 on: December 17, 2018, 06:36:00 am »
I also see it only in the 30V range. The measurement above with +/- 0.4ppm was done in the 3V range.

I've done some more runs this night. The "switching" points always appear at the same voltages.
 


Share me