Author Topic: Metrology Meeting in Stuttgart/Germany  (Read 70640 times)

0 Members and 1 Guest are viewing this topic.

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5468
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #75 on: May 26, 2019, 01:44:36 pm »
Is anyone interested in low ohm resistance comparison?
I could bring my collection of Burster Kalibrierwiderstaende and the fitting cables for the Keithley 2450 SMU and 3458A

0.001 Ohm to 10 Ohm


There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4522
  • Country: de
  • Per aspera ad astra
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #76 on: May 26, 2019, 07:02:22 pm »
Today I got confirmation of another guest present on the event. Let yourself be surprised.

Kleinstein will show up and do a speech?  ;)   :-+
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14159
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #77 on: May 26, 2019, 07:17:50 pm »
I will not be there. So no speech and presentation of DIY  high resolution multi-slope ADC.
I though about coming, but did not find the time.
 
The following users thanked this post: The Soulman

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #78 on: May 26, 2019, 07:54:54 pm »
Hello,

from Flinstone I got the following mail:
(see also 2nd entry on branadics list)
https://www.eevblog.com/forum/metrology/volt-nut-meeting-2019-in-stuttgartgermany/msg2426568/#msg2426568

Quote
I will bring with me:

 

    PTFE cable spools                             2C+S and 4C+S   (white PTFE Teflon with 0.91mm silver plating C = conductor, S=braid/shield)
    Au plated spade connectors        suitable for 3458A/8508A/8558A/8588A/etc. … (Au plating on top of pure copper)
    LTZ1000ACH                                       sourced from ARROW – new and unused parts
    2DW232                                               Correct brand and date code
    Au plated banana plugs                 Staubli 4mm (6 colours)

 

T&M equipment:

    METRON/WAVETEK/FLUKE 7000 series “M69”    (10.000 041 ± 2.5*10E-6)V           last Cal performed by JRP using traceable equipment from NPL London UK 8 May 2019
    RS-2       10kΩ Standard reference resistor (with EXTECH/FLIR RHT sensor)              Its value is well known up to 7 digits and ±0.5ppm uncertainty.
    Keithley 7510 or DAQ6510 to monitor RS-2 temperature inside enclosure
    Laptop running software to monitor WAVETEK with a 3458A unit … - I hope we can get assistance from someone to share his 3458A unit during afternoon Cal sessions.

(I cannot take my 3458A 2ppm unit as it is too heavy and vulnerable, I have no insurance in case the unit is damaged during transportation to Stuttgart).



with best regards

Andreas
« Last Edit: May 27, 2019, 06:02:36 pm by Andreas »
 
The following users thanked this post: The Soulman

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #79 on: May 27, 2019, 07:55:07 am »
I got a similar pm from a member where i guess, that could be flintstone.  :-X
 

Offline branadicTopic starter

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #80 on: May 27, 2019, 12:06:11 pm »
I'm sorry for all the late decider, but the time slot for registration is over.
To all others joining the event, you can bring your equipment on friday til 5:00pm to the event location to install it in the lab.

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 

Offline guenthert

  • Frequent Contributor
  • **
  • Posts: 712
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #81 on: May 27, 2019, 04:46:25 pm »
I will not be there. So no speech and presentation of DIY  high resolution multi-slope ADC.
[..]
Well, you could present it here, on-line  ^-^
 

Offline branadicTopic starter

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #82 on: May 27, 2019, 05:47:43 pm »
I will not be there. So no speech and presentation of DIY  high resolution multi-slope ADC.
I though about coming, but did not find the time.

What a shame, hope you will find the time at the next event including a speech, I'm sure there are a lot of people interested in your project. Until then, a lot of time to finish the design :)

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 

Offline babysitter

  • Frequent Contributor
  • **
  • Posts: 893
  • Country: de
  • pushing silicon at work
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #83 on: May 27, 2019, 06:24:21 pm »
Glad to hear about the intentions to have a "next time" (then without changing the SI definitions, I guess.)


I'm not a feature, I'm a bug! ARC DG3HDA
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #84 on: May 28, 2019, 05:12:40 pm »
Currently we think which Hotel is the most meaningful to book...

In which Hotel we will meet the most of the attendees? Could somebody who has already booked give us a hint?

Thx
Philipp
 

Offline branadicTopic starter

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #85 on: May 28, 2019, 05:29:33 pm »
I suggest Campus Guest, which is 10min walk to the event, 5-10min from the next underground station and 3 stations into the center of Stuttgart.

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: e61_phil

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5468
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #86 on: May 28, 2019, 06:07:56 pm »
I am booked at Campus Guest by email and they said they had plenty of rooms available.
There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: e61_phil

Offline branadicTopic starter

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #87 on: May 28, 2019, 09:35:51 pm »
I will have my temperature chamber with me, so if someone wants to perform a temperature cycling, you will have the chance at the event. It takes about 2h from room temperature to 15°C, all the way up to 45°C and back to room temperature.

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #88 on: May 31, 2019, 09:02:48 am »
Hello,

I think I have asked that before somewhere but the elementary question is:

if a device has been calibrated before 20.05.2019:
how much do we have to add / subtract to get the "real" value.

with best regards

Andreas
 

Offline MiDi

  • Frequent Contributor
  • **
  • Posts: 600
  • Country: ua
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #89 on: May 31, 2019, 02:54:54 pm »
Hello,

I think I have asked that before somewhere but the elementary question is:

if a device has been calibrated before 20.05.2019:
how much do we have to add / subtract to get the "real" value.

According to mzzj it is 0.107 ppm. For 10V this is 1.07uV.

Edit: Clarification
« Last Edit: May 31, 2019, 08:51:54 pm by MiDi »
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #90 on: May 31, 2019, 03:08:35 pm »
Thanks,

from the linked pdf of fluke:

Quote

The new value for the Josephson
constant (2e/h) is slightly smaller than the
1990 value, so a device that is measured to the
2019 value will be larger than the 1990 value
by approximately 0.1 ppm.


So as a volt-nut we have to add 1uV on the before 20.05. calibration readings of a 10V reference.

with best regards

Andreas

 

Offline guenthert

  • Frequent Contributor
  • **
  • Posts: 712
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #91 on: May 31, 2019, 04:35:05 pm »
You're not serious, are you?  Comparing two potentials at 10V above ground to less than 1uV is challenging, but within reach of the (very) dedicated hobbyist.  Measuring those potentials with an accuracy of less than .1ppm so that historical values can be compared is out of the question.
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #92 on: May 31, 2019, 04:44:01 pm »
You're not serious, are you?  Comparing two potentials at 10V above ground to less than 1uV is challenging, but within reach of the (very) dedicated hobbyist.  Measuring those potentials with an accuracy of less than .1ppm so that historical values can be compared is out of the question.

He just said add 1uV to the calibration value. That seems to be perfectly fair.  :-//
 

Offline branadicTopic starter

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #93 on: May 31, 2019, 05:40:34 pm »
Metrology Meeting 2019 already started today. A few members already received, their items are installed in the temperature controlled lab, first discussion started and we will have dinner in 20min.  :-+

#Volt-Nut

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: Andreas, hwj-d, The Soulman, 2N3055

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #94 on: June 01, 2019, 01:31:53 am »
Somebody should do a livestream  >:D
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline The Soulman

  • Frequent Contributor
  • **
  • Posts: 949
  • Country: nl
  • The sky is the limit!
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #95 on: June 01, 2019, 12:44:49 pm »
Someone should post some pictures and data.  :-/O
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #96 on: June 01, 2019, 06:47:01 pm »
Hello,

what do you expect?
-lots of people staring at 8.5 digit displays looking wether the last digit will flip (or not)?

BTW: look at the calibration date of the lower 3458A: it is within 24 hrs specs!!!

with best regards

Andreas
 

Offline The Soulman

  • Frequent Contributor
  • **
  • Posts: 949
  • Country: nl
  • The sky is the limit!
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #97 on: June 01, 2019, 07:00:49 pm »
Hello,

what do you expect?
-lots of people staring at 8.5 digit displays looking wether the last digit will flip (or not)?

Yes! exactly that.  :-+

BTW: look at the calibration date of the lower 3458A: it is within 24 hrs specs!!!

with best regards

Andreas

Is the source that good?
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #98 on: June 01, 2019, 07:07:57 pm »
Its the Fluke 7000 from above.
So 10.000 041 +/- 2.5 uV

Have not looked up the calibration report of the 3458A up to now if the value fits to the calibration offset.

with best regards

Andreas
 
The following users thanked this post: The Soulman

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3238
  • Country: de
Re: Metrology Meeting 2019 in Stuttgart/Germany
« Reply #99 on: June 01, 2019, 08:16:42 pm »
From calibration protocol (of yesterday):

for +10V the 3458A reading was 3.4 uV too high against calibrator actual value.
for -10V the 3458A reading was 4.1 uV (amount too high)
Calibration uncertainity is 2.5 ppm (+/-25uV) with 24 hrs we get 2.6 ppm (+/-26uV)

so for 10V + 56.8uV we have to correct to 10V + 53.4 uV
for -10V -58.5uV we have to correct to -10V - 54.4uV

the difference to 10V + 41uV from Fluke 7000 is 12.4 or 13.4 uV which is within the +/- 26 uV +/- 2.5 uV uncertainity.
But which device is showing the truth? -> we need more references/DMMs   :-//

Edit: what where the results between the Fluke 7000 and the other Flukes (732)?

Edit: And even more evil: I did not find in the calibration protocol if the calibrator has been already set for the new volt  :-DD

with best regards

Andreas




« Last Edit: June 01, 2019, 08:40:32 pm by Andreas »
 
The following users thanked this post: The Soulman


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo