IMO it is good and simple to keep TUR 4:1, however in real world it is very challenging, sometimes you get TUR 2:1 and that is guardbanding come into play.
I'd rather not to see mention of TUR anywhere in my 17025 calibration certificates as there seem to be about 50% probability that cal cert doesn't fullfill ISO17025 requirements in that case.

TUR is a huge pain in the ass if you are dealing in the 17025 "ecosystem", if the calibration certificate doesn't include total calibration uncertainty then its a guesswork to calculate it from incomplete data and entire idea of traceability is questionable.
The 0.03cel calibration uncertainty(k=2) for 0.1cel resolution fridge thermometer example was from one of these.
BIG difference between 17025 and "TUR calibrations" is that the the uncertainty calculation per ISO 17025 or EA4/02 needs to include also the uncertainty contribution of the calibrated device(DUT).
If the DUT resolution is 0.1 volts it already makes 0.1/sqrt(3) = 0.06 volt standard uncertainty and 0.12 volt expanded (k=2) uncertainty
If the DUT reading last digit is jumping around ie +-0.1 that would add another 0.2/2/sqrt(3) to the calibration uncertainty.
So claiming 0.03 volt calibration uncertainty (k=2) would be very suspicious if the meter resolution is only 0.1 volts (or whatever units)