Bonjour a tous!
After a long pause for travel, work, medical,  back to the VoltNuts....
Voila!  a 10.000 V ref box.  power ext 18V lab supply or  or 2  9V internal batteries.
 Adapted a German CircuitValley.com  LM399 board, inside cast Pomona box,
Added  15V  prereg 15V LM2940-15 LDO   3 way power switch ,  1N5818 Schottky polarity protect diodes,  LED and binding posts.
The  LM399 has a Styrofoam heat shield , removed for photos.
Shown with Keysight 34465A  1 yr, in CAL.    9.999968 V  after warmup.
Can replace the LT1001 ACNS easily as it is on gold-plated machine pin socket.
Seeking an independent CAL of the board.
Your thoughts greatly appreciated!
Bon Soiree,
Jon
The Internet Dinosaur..
passionate about analog electronics since 1950s
Offline Andreas
Re: 10V ref LM399 project
« Reply #1 on: January 30, 2023, 07:57:03 am »
Hello,
just some questions/notes:
- where is the 15V voltage regulator. I hope it is after the schottky diode at least for the mains input. There is a output voltage dependancy of the LM399 on heater voltage.
- Why do you use a FET with only +/-12V max gate voltage for a input voltage of ~18V?
- Why a Transient zener with 30 V (usually max operating voltage) which begins to conduct at ~40V
- Since a Transient zener usually creates a short on overloading I would put a fuse before the zener.
- The Schottky diodes at the output are unclear for me (perhaps the drawing is hard to read).
- I hope the trim range is very small for the output voltage (usually trimmers have rather high tempco).
  usually a different trimming scheme (similar to standard cell replacement in the data sheet) is used for precision trimming.
- before calibration I would at least do some run in phase for 2 kHrs
- and of course you should do some tempco tests for the whole cirquit
- the housing needs some markings (input voltage, output voltage and orientation of the package).
  the output voltage is dependant on orientation of package.
  I had some strange drift after putting the LM399 up side down accidently for some days after a calibration
  I had luck that it reversed after some weeks in normal operation.
with best regards
Andreas
Online Kleinstein
« Reply #2 on: January 30, 2023, 08:35:32 am »
The trim circuit is not ideal. A better place would be in parallel to the other resistor in the gain setting part and than with a series resistor for the coarse part.
The LED should get away with less than 10 mA off current - this helps with battery operation. Modern LED should get away with ~ 0.2-1 mA.
The LT1001 is a bit power hungy with some 2 mA typical. For battery operation I would consider a lower power OP-amp, like OPA207 (SMD only and thus not that practical for the existing PCB).
For judging on the cause of drift, it would help to also have acess to the raw 7 V ref voltage.
I personally would prefer a different connector for the supply and if wanted an adapter cable to connect to a lab supply if needed.
Offline iMo
  • Country: nu
« Reply #3 on: January 30, 2023, 08:41:47 am »
The major issue I see with the RoadRunner's reference board is the "zener" is not bootstrapped off the opamp's output (a "standard" today). It hangs on the Vcc (not stabilized) and the opamp makes 7->10V only.
Offline r6502
« Reply #4 on: January 31, 2023, 11:49:26 pm »
Hi Jon,
If you have it not already seen, have a look here:
LM399 based 10 V reference
May be, you are also interested in the new ADI ADR1399, then have a look here:
ADR1399 reference
Lots of useful info out there.
Guido