Hello,
I am looking for the schematics of the serial interface T1678 between
 the wiltron 68XXX synthesizer and a PC in hyper terminal mode.
Thank you

acbern:
I guess you need this to do the calibration steps mentioned in the manual? I heard (from a person within Anritsu) that these units somtimes loose lock, and when you do these adjustment steps over serial, these errors go away. Any experience re. this?

Anyway, mine is working, but I wanted to be prepred, and was looking for the same data as you, no success. So all I found out was the following:
The RJ connector on the back supports 2 serial lines. The T1678 seems to do nothing else than splitting these up into two separate connectors.
I have done some measurements on the connector, and my conclusion was that, when you look at the connector with the locking nut in upward direction, so pins on the lower side, the two left pins are RxD, the two right pins are TxD. This is given without a guarantee, but it should alow to build up a test cable and figure out the rest. So in other words, which signal relates to serial 1 and which to serial 2, I have not figured out, I stopped there.
It would be nice if you could publish your results here, for teh sake of the community.

Aiy:
Hello Everyone,

Sorry for digging up such an old thread, but I found some possible info on the RJ45 connector in a PDF and wanted to post it since it was hard to find.
The PDF is titled:

Wiltron
Option 10
Complex Modulation Capability
for the Series
682XXB, 683XXB, 692XXA, and 693XXA
Synthesized Signal Generators

https://optech.kz/f/rukovodstvo_polzovatelya_4.pdf

See attached image for pinout.

Best regards
AIY

Aiy:
Hello Everyone,

I just wanted to post my findings with the RJ45 connector on the Anritsu 68/69000 synthesizers.
I cut and soldered a dual DB9 UART according to the cable pinout today and it worked.
The primary UART is 19200 baud, 8, n, 1, Xon/Xoff.
The secondary UART is 9600 baud, 8, n, 1, Xon/Xoff.