It was a gamble buying this instrument in a broken condition. The fault was described as:
This is a PARTS UNIT. Sold AS-IS. When you turn on the Output (Voltage or Current) it immediately goes to -36V.I couldn't find a service manual or schematics on the internet for this or similar Yokogawa instruments. So I knew I was going to have fix this without a schematic.
When I confirmed that the output was at -36V on all the ranges.
I traced out the power amplifier section, expecting to find a shorted transistor. All the transistors turned out to be good.
I then found the op-amp that was the driver for the output stage. If I shorted it output to ground the output also went to 0V. Indicating that the output amplifier was not the problem.
I determined that the op-amp, an LT1012, was damaged. Initially I replace the op-amp with an OP07. This is a similar op-amp but not as good.
This didn't fix the problem.
I looked for more op-amps that weren't working properly. I found a damaged NE5532 op-amp that was also not working. When I change this the unit started working properly.
These are the two amplifiers that I changed.
Regards,
Jay_Diddy_B