PE1RQF:
Apply rated signal at 60Hz is correct. it is the filter adjustment.

pasta2018:
now I'm a little confused. As in the picture I uploaded, for
10V,apply 10V 60Hz
15V, apply 15V 60Hz
30V,apply 30V 60Hz
60V apply 60V 60Hz
100V apply 100V  60Hz
150V apply 150V 60Hz
100V(500) apply 100V 500Hz
100V (5.5k) apply 100V 5.5kHz
Are these right?

dl1640:
That's correct.
(500Hz), (5.5kHz) for measurement under LPF set to ON.

The reference signal applied is at commercial frequency 60Hz/50Hz will be fine.

pasta2018:
ok, I got it. Thanks very much.