YOKOGAWA WT1600 Power meter calibration mode
pasta2018:
Hi, I 've got a wt1600 repaired, changed the sample resistance in the sample board, now I test It and find the error is out of tolerance. I have got Fluke calibrator, but I don't know how to get in the calibration mode of wt1600, can some one provide a service manual of tell me how to get in the calibration mode, many thanks.
pasta2018:
there is a adjustment procedure for wt230, maybe I can try this method.
pasta2018:
That doesn't work on a wt1600, any other suggestions?
PE1RQF:
Hi,
Try switch on instrument while pressing display button. After start up, press shift and than display button.
Regards,
Erik
pasta2018:
It's really great, that works. Really thanks.
I will try to calibrate it.