YOKOGAWA WT1600 Power meter calibration mode

(1/3) > >>

pasta2018:
Hi, I 've got a wt1600 repaired, changed the sample resistance  in the sample board, now I test It and find the error is out of tolerance. I have got Fluke calibrator, but I don't know how to get in the calibration mode of wt1600, can some one provide a service manual of tell me how to get in the calibration mode, many thanks.

pasta2018:
there is a adjustment procedure for wt230, maybe I can try this method.

pasta2018:
That doesn't work on a wt1600, any other suggestions?

PE1RQF:
Hi,
Try switch on instrument while pressing display button. After start up, press shift and than display button.

Regards,
Erik

pasta2018:
It's really great, that works. Really thanks.