Author Topic: Zero Ohm connector plates ? an advantage compare to single point ?  (Read 1264 times)

0 Members and 1 Guest are viewing this topic.

Online OverspeedTopic starter

  • Frequent Contributor
  • **
  • Posts: 482
  • Country: fr
Hello

In NIST 1458 Measurement Service for DC Standard Resistors , they use a triangle zero ohm rather than a single point connection

Does that really provide an gain / improvement as that not complicated to machine from copper plate .

Regards
OS
 

Offline Wazza

  • Contributor
  • Posts: 19
  • Country: au
Re: Zero Ohm connector plates ? an advantage compare to single point ?
« Reply #1 on: October 15, 2024, 07:40:56 pm »
The 'advantage' when presenting  zero ohms for a *4 wire* measurement is that the current flow from the stimulus/source terminals does not pass through the meter terminals in any configuration. This reduces possible measurement errors.

Of course thermal emf errors could still be present, depending on the materials used to make the connector plate; but that's another rabbit hole to fall down....

Waz
 
The following users thanked this post: Victorman222, Overspeed

Offline manganin

  • Regular Contributor
  • *
  • Posts: 196
  • Country: fi
Re: Zero Ohm connector plates ? an advantage compare to single point ?
« Reply #2 on: October 15, 2024, 09:34:48 pm »

See section III in the SR1010 manual:

https://www.ietlabs.com/pdf/Manuals/SR-1010_im.pdf

 

Offline thermistor-guy

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: au
Re: Zero Ohm connector plates ? an advantage compare to single point ?
« Reply #3 on: October 16, 2024, 12:28:06 am »
Hello
... they use a triangle zero ohm rather than a single point connection

Also known as a four-terminal junction. It can be circular. See fig. 12 in Riley 1967
 
The following users thanked this post: Overspeed

Online OverspeedTopic starter

  • Frequent Contributor
  • **
  • Posts: 482
  • Country: fr
Re: Zero Ohm connector plates ? an advantage compare to single point ?
« Reply #4 on: October 17, 2024, 11:27:20 am »
The 'advantage' when presenting  zero ohms for a *4 wire* measurement is that the current flow from the stimulus/source terminals does not pass through the meter terminals in any configuration. This reduces possible measurement errors.

Of course thermal emf errors could still be present, depending on the materials used to make the connector plate; but that's another rabbit hole to fall down....

Waz

Hello

Thanks for the links

Thermal emf is a concern in design as that mostly impossible to avoid material mix , but try to be close is possible .

Can be interesting to design a mechanical ''frame'' or old style circuit and make tests .

Regards
OS
 