What kind of messages are displayed?
I know the normal one, greetings text and partial modules found text, starting with C.
What kind ofs are error messages?
Is text "Error" displayes and so on.
Somewhere is said that '?' is displayed, what it look like?
When is sign in front of exponent displayed, is it with exponent only?
What device specific messages are outputted?
I know the one with "8505 :[modules]" but are there others.
There are actually very few textual messages. I have edited the manual's explanations so they are easier to understand if you don't have the hardware:
1) There is the "HI-5.0.6" version number at start, that goes away after a short time. (around one second, maybe less)
2) Then there is the installed module scan. If a module is missing, its corresponding number will be replaced by a space in the array of presumably hex digits. This also remains displayed for about one second.
3) Then there are some "momentary error messages" that are recoverable without a reset:
Error 0
Attempt to set zero in wrong function (not V DC or OHMS) or an overrange has been entered.
Error 1
"Store" function attempted during overload condition. Change to higher range or (if storing cal correction factors) use lower value source.
Error 6
Display overflow. Check offset and scaling values.
Error 7
External Reference - voltage on one input exceeds 20V dc.
Error 8
Controller module is faulty. Replace controller module.
Error b
Illegal push button sequence in Calibration mode.
Error C
Invalid push-button sequence, or illegal value entered.
Error d
Calibration Memory faulty or not installed. Occurs when storing into, or recalling from, Calibration Memory, or at power-up. Replace or install Calibration Memory chip.
Error F
Cal Memory check-sum problem. Restart the meter. If necessary, reprogram Cal Memory. Replacement of Calibration Memory may be required.
Error H
Ohms connection problem. Verify that all connections are proper. Check input fuses. Check input lead polarity in four terminal connections.
4) Finally there are the "latching error messages" that are unrecoverable, the meter has to be powered down:
Error
(No error ID) System error, usually appears at power-up or reset. Repeat power-up or reset.
Error 2
Filter module faulty or not installed.
Error 3
DC Signal Conditioner module faulty or not installed.
Error 4
OHMS, A DC, or A AC error. Applicable module may be faulty or not installed. Check for bad input level.
Error 5
Analog to Digital Converter Module error.
Error 9
Function selection error. The function module selected is faulty or not installed. V DC problem causes Error 3.
Error E
More than one type of ac converter, or the wrong type of ac converter for the instrument, is installed. Also appears when a Calibration Memory module from an 8502A is installed (it should not be installed in 8505A 0r 8506A which have calibration EPROM in the controller itself)