Ok before I do my question an introduction story will be the below text .
Text
The conductance function in the meter reads out as a value of nanosiemens.
This range allows to measure high resistance values from 20 Megohms to 100,000 megohms.
The formula for conversion between nS and megohms is Mohm=1000/nS
so for example 10.00 nS=100 megohms.
Conductance is another way to measure high resistances, such as leakages in diodes, capacitors, pcbs, or insulators or even motor winding insulation or leakage.
The main difference between a megohmeter and the conductance test is the conductance test, uses only a few volts to perform the test where the megohm meter can be set to much higher voltages for "stress" tests.
When available stress testing is a better method especially if the insulation being tested is marginal but stress testing electronic components or circuit boards at higher voltage may cause damage so the conductance tests may be more appropriate in these cases.
If the insulation was completely deteriorated the conductance method should show bad results as well.
INSULATION TESTING
Insulations above 50 megohms is generally considered good.
Insulations between 1 megohm and 50 megohms are questionable and should be check frequently as a failure may be about to occur. Insulations below 1 megohm are considered bad.
Multimeters in either the ohms mode or the conductance mode (nano Siemens mode), only use about a volt which does not stress the insulation.
CONDUCTANCE MODE
A Siemen is the recipical of an ohm. If you are going to make resistance measurements over 500 megohms, then the 50.00nS conductance range will allow you to read up to 100,000 megohms. To convert the readings into ohms, you need to recipicate the Siemen reading. Note that nano is 10 to the minus 9. Here are a few values:
0.01nS = 100,000 megohms
0.1nS = 10,000 megohms
1nS = 1000 megohms
10nS = 100 megohms
50nS = 20 megohms
60nS=16.6 megohms (Fluke 28II limit down)
90nS = 11.11 megohms
300nS= 3.33 megohms (Agilent U1272A limit down)
And now my question :
I am trying to measure the high impedance resistor of my DMM Fluke 28II and Agilent U1272A.
And looks impossible.
The Fluke 28II at the Ns mode gets to 0.01Ns with out probes connected,
The U1272A shows 00.04 with no probes connected,
and the Fluke 8050A shows about 00.08 with no probes connected.
I am trying to use relative, but I do not get any reading at all or equal to zero 00.00 Ns
Why Is so hard to find out the High impedance resistor in those DMM ?