PN Voltage Resistance Leakage
TLP3558A - 160V 2R 1uA $1.5-2.50
AQY212EH - 60V 2.5R 1uA $0.25+
AQY210EH - 350V 25R 1uA $0.25+
AQV253A - 250V 8R 1uA $0.8
LH1540 - 350V 22R 1uA $0.8
PVA1354N - 100V 5R 1nA? $2+ (limited life)
PVA3324 - 300V 24R 1nA? $2
- Equivalents for AQY: TLP227G, TLP222G, G3VM-351D, LCA110L (6pin, lower R typ) $0.5-3
Thanks for the notification, this is really good documentation :-+. Hope to build it.
Some photoMOS, first is one OP used:
- TLP3558A - 160 V 2 R - $1.5-2.50
- AQY212EH - 60V 2.5R - $0.25+
- AQY210EH - 350 V 25 R - $0.25+
- Equivalents: TLP227G, TLP222G, G3VM-351D, LCA110L (6pin, lower R typ) $0.5-3
Not sure how much resistance matters? Shame if 25R is too high as there are a ton of surplus ones, probably from telecom gear.
Voltage will depend on what circuits you plan to measure. Original spec was ~200V but I'm sure the relays were rated for slightly higher.
I would expect the 25 Ohms or even 100 Ohms to be still OK. It may effect the 10 MOhms input impedance ranges a little, but this would be higher than 10 V and usually the lower accuracy ranges anyway.
I would be more worried with leakage between the channels. So it may be a good idea to not use the very low resistance ones. At least unused channels should not contribute much to leakage. So the scanner may not be a good choice for high impedance source.
One thing I'm not sure of, what is the purpose of channel 10 on the original (https://github.com/cozdas/CozScan2020/blob/master/Documents/KeithleySchematic.png)scan card? It uses K111 relay and U105 3301 photomos in series, with some 100R resistors. The other channels use a single relay.
//The SPI clock line from K2000 idles high, whilst from DMM6500 it idles low.
//Determine idle state of SPI clock by waiting for 50 samples of the same value