So, I tested a couple of capacitors using the method above (soak at 10.00000V overnight, disconnect the supply in the morning and let sit for an hour, take the first measurement, let capacitors sit on the bench with nothing connected for 9 hours or so, then take a second measurement).
C1 - Large Mica capacitor, rated 0.25uF at 100V
Start Voltage at 08:39 was 9.71413V
Stop Voltage at 17:49 was 8.04941V
Delta: 9.16h, 1.66472V
Lost Q: 0.41618E-6 Coulomb
Avg Leakage current: 12.6pA
C2 - Polycarbonate capacitor, rated 5uF at 50V
Start Voltage at 08:40 was 9.95304V
Stop Voltage at 17:50 was 9.79666V
Delta: 9.16h, 0.15638V
Lost Q: 0.7819E-6 Coulomb
Avg Leakage current: 23.7pA
Do the numbers seem reasonable, or did I mess up the math...