Hi all!
I guess this is a general issue on many instruments, but I'd like to add a voltage output to my RM5/1 ratemeter, to log the count rate as a dc voltage either with an arduino or a chart recorder (just to keep it 1970's).
-I can get to the meter terminals
-The meter is driven by a current (I assume) so I guess whatever I add should be very high impedance as to not alter the meter reading
-Current needs to be converted to a convenient voltage (say 0-100 mV or 0-1V)
-No schematic on the ratemeter, it's all transistors. I'd rather not mess too much with the base board.
-Instrument runs on 3V.
[update] Meter is a Crompton instruments 200uA F.S.D. one
Is there a neat, simple way of doing this? This MUST have been a common mod to do to instruments in the 1960's and 1970's right?
Thanks in advance!
If it is a plain old moving coil meter (PMMC), it will be quite low in impedance.
You can see if it is affected by a fairly high parallel resistance/impedance as follows:
(1) Set the instrument up to give a convenient meter reading.
(2)Connect a DMM across the PMMC terminals, on DC volts(the former may well autoscale to mv while you are doing this test) manually record the both the DMM reading & that of the PMMC.
(3)Whilst still maintaining the same input to the instrument, disconnect the DMM & see if the PMMC reading increases.
If there is no discernible change, you know that 10-11 Mohms are not a problem, & can design your connected circuit with such an input impedance.
If you think this is a excessively high value of Z, try hooking progressively lower values of resistor across the PMMC, until there is a discernible drop in its reading.
You then know your lower limit for Z.