Author Topic: ADS1248 - surprising noise versus Vin graph  (Read 1248 times)
Online splinTopic starter
« on: February 17, 2019, 07:34:17 pm »
I just came across this graph in the ADS1246/7/8 datasheet (24 bit SD ADC) which rather surprised me:
Noise v input level graphs aren't especially common in ADC datasheets, but I've never seen one that looks quite like this!
The noise level varies dramatically with Vin, between 50nV and 270nV. Typically I would have expected the noise to be relatively flat v Vin, perhaps rising towards the maximum i/p levels where the reference voltage noise starts to become a bigger percentage of the total. (I've seen one where the noise actually reduces at the maximum input, suggesting the reference noise is negligable compared to ADC noise).
Unfortunately there is nothing in the datasheet referencing this graph or how the measurements were taken, other than Vdd and PGA gain shown on the graph itself. In particular it doesn't say if the internal, or an external reference was used and perhaps more importantly how long the noise was measured for at each input level -  assuming it was done discretely rather than using an analogue sweep (with unspecified noise bandwidth).
So would you be happy if you'd selected this ADC on the basis of the 70nV rms noise stated in the noise performance table, not having spotted this graph? Note that the noise performance tables published in pretty much all ADC datasheets are with the input shorted (ie. 0V), and there is no reason to doubt the 70nV figure, but given the unusual wild variation with i/p level shouldn't that have been clearly stated somewhere - eg. by referencing this figure 9 as well as the noise tables in the noise specs?
Is this noise characteristic peculiar to this particular ADC (I hope so) or might there be other, similar third-order (or higher) sigma-delta ADCs with similar, noise characteristics but unspecified in the datasheet?
Is the graph simply wrong? It's not uncommon to have nonsense graphs in datasheets.
Offline Kleinstein
Re: ADS1248 - surprising noise versus Vin graph
« Reply #1 on: February 17, 2019, 08:35:07 pm »
The curve looks like it comes from a measurement of a single sample. We don't know how the measurement is done. So there could be some uncertainty from the measurement. So I am also not so sure one can take the curve for accurate in every detail. Even if reproducible for the given unit, chances are a different unit would behave different.
Sigma delta ADCs are known to have some odd behavior at some voltage levels. This is known as idle tones and can lead to extra noise at some constant voltage levels. So chances are a odd looking curve for the noise is not that unusual for SD ADCs. The details of the higher grade SD ADCs are different and partially well kept secrets and much of this about avoiding the idle tones. With many different solutions the different ADCs may behave quite different.
AFAIK idle tones are a problem only for constant input levels. So in some applications it may not matter.
