Well it seems things just took a turn to the worse... I've noticed some issues with a few of the range buttons so I tried to apply some contact spray I had around for a while and never got around to use it. I applied a generous amount and made sure the spray goes into the switches. I let the whole thing dry for 1 hour and closed up the meter, now I am getting "1" on all parameters on all ranges

This makes me very sad... so I will open it again and check the voltages 1st and see from there. I could not see a spec. on the voltage rails when I come to think about it...
Update - I connected the meter to power again for taking voltage readings on the rails. I tested the DC voltage and it won't read anything at all on all ranges. On the resistance range I can get 8.825k (not very stable, the LSB keeps on drifting down slowly) but the resistor is (measured) 7.5k so something is not OK.
Test results:
Ohm range / Ohm display
200 -> 1.45
2k -> .0012
20k -> 0.001
200k -> 00.00
2000k -> 000.0
20M -> 0.000
Voltage Range / Voltage input / Voltage display
200m -> 188m -> 84.41m
2 -> 1.984 -> 0.0843
20 -> 19.34 -> +1
200 -> 31.91 -> +1 HV
2000 -> 31.91 -> +1HV
That's what I've tested so far. Does not look good.
Oh, and voltages:
TP2 (13V) - Started at 13.0V when I powered up and than started to get more and more noisy and now drifts from 12.2 to 13.1V. Does not seem very DC to me so I connected my old scope and took a picture. See TP2.jpg The period is about 11 or 12 ms so that's around 83 to 90 Hz.
TP3 (6v) - Around 6.14V but drifts up and down, scope image is the same as TP2.
TP4 (-5V) - Around -5V, same drifts and AC as above.
TP5 (-10V) - Around -11V to -12V. Same drifts and AC as above.
The AC on the said test points is about 5.1V ppk. Should I swap the power input filtering caps 1st?