All,
Got a bit of a puzzler for ya all, and while typing out a post about it being burden voltage on a meter, I realized it may actually be the resistance change of the meter measuring current. So, here we go. And maybe this belongs in the test equipment section, but I figured this was more a question regarding Burden Voltage rather than the test equipment specifically.
Was measuring voltage/current/power draw of an iPhone over USB. Using a simple inline USB Tester (
https://www.tindie.com/products/FriedCircuits/usb-tester-20/). Setup was two Keithley 196's, one for voltage, one for current. Both set to auto-range.
So, when plugged in at first, iPhone would draw about .75A, then as the SoC of the iPhone batt would increase, the draw would lessen. Eventually, the draw got to below .3A, and the meter would autorange to milliamp range, and the phone charge would stop saying "unsupported accessory". This could be fixed by turning off auto-range, and dropping it back to the 3A range.
Now, my thought is that this was burden voltage, but according to the Principals of Operation for the Keithley 196, and I quote:
The resistor current shunt network R28 is configured so that a full scale current input will result in a 300mV drop across the network on all current ranges. For DCA, this voltage is routed to the multiplexer through analog switch U24B. For ACA, the signal is routed through U21D X10 amplifier U28B. The amplifier signal then travels through analog switch U2lB and buffer U26B to the TRMS converter. The converted DC signal is then routed to the multiplexer.
So, I'm a bit confused by this. It sounds like there should be no real change due to the range changeover, but obviously there is. What's the deal?