The el cheapo LCR tracerHi, I wanted to share this little project of mine.
I have made a board and written some code to use an oscilloscope and a signal generator as an LCR meter, this is nothing new, I know there are multiple tutorials to make so, but they often require manual work writing down values in an Excel sheet and making the calculation, I have not time for that! What I have instead is a bunch of GPIB instruments, so I can offload the work to them.
The start is this app note from Tektronix, again nothing new, but it is the example I have used as it seemed the easier one to automatise, it uses the I-V method to measure impedance.
https://www.tek.com/en/documents/application-note/capacitance-and-inductance-measurements-using-oscilloscope-and-function-geThe requirement, if you wish to make your own, are:
a signal generator of decent quality that you can control, so either with GPIB or serial. Flatness is not a requirement as the voltage is measured anyway, low DC offset and low AC leakage are preferable.
And an oscilloscope, again controllable from a pc, should have built-in phase measurements between channels,
If you decide to use my pcb design be aware that the BNC spacing on my scope is 30mm, your scope might be different. You can either change the design or replace the female connectors with male ones and run cables between the
I’m using a Agilent 54624A
I have tried a few such as HP 8904A and Philips PM5193, the main difference is going to be the maximum frequency they can provide, for my use 1 MHz maximum frequency is more than enough, and as I don’t understand how to do impedance control in this context it probably better to stop (well) below 10 MHz.
The hardware My board has a few changes from what you will read in the Tek app note:
- I’m using 4 precision 1k resistors rather than 1 to experiment with ranging (1000 , 500 and 250ohm), you might not need to populate all of them. 1000 and 500 ohm are the more useful ranges, I fear 250 is getting too close to the 50ohm output impedance of the signal generator and might be an issue.
- The reason for ranging is that with 1k you often get a very small voltage on channel 2 and it might be a struggle to measure it.
- A 50ohm resistor is added to the input to terminate the connection to the signal generator.
- A dc booking capacitor is added as an option to either clean up a misbehaving signal generator or have a separate DC source.
- Finally, the Tek note describes using 10x probes, I cannot do that with my scope, the signal on channel 2 is often too small to be measured this way. Other scopes might give you more freedom in this regard and then a 10x probe would load the circuit less.
PartsBNC connectors
https://it.aliexpress.com/item/32850462301.htmlSMA connectors
https://it.aliexpress.com/item/32248312829.htmlPrecision 1k resistor that doesn’t cost too much TE ConnectivityYR1B1K0CC
I repurposed a SMA cable to make the probes
And you might need a SMA to BNC cable to connect to a signal generation
ScriptThe code example that I can provide use PyMeasure
https://pymeasure.readthedocs.io/en/latest/ and an AR488 Arduino-based GPIB adapter. Other adapters and serial interfaces are also supported by PyMeasure.
It should be easy to adapt the code to different instruments, the more problematic area is the auto-ranging. Auto-ranging is needed to keep the signal at a good scale and use as much vertical resolution as possible, different scopes will handle clipping in different ways, so you will have to figure out how to adapt the procedure.
At the moment I have two main modes to capture data, swipes, to make nice plots over time of ESR inductance and capacitance or single measurements (more similar to other LCR meters). A single measurement does L C and R measurements at fixed frequencies, 100hz or 1khz for C and 100khz for ESR are some common standards, useful to compare with data sheets, but you should choose what you need accordingly to the context.
Questions There are a few things to improve and I would like to ask for some support.
At ht moment there is no provision for calibrating the systems apart from calibrating the separate instruments.
I would like to add a open-short calibration procedure,
As I understand a short calibration would create a baseline for ESR, I’m not sure how to make a baseline or correction for capacitance.
Even further I would imagine adding a frontend such as something like this
https://www.ti.com/lit/ug/tidueu6b/tidueu6b.pdf?ts=1685491422213&ref_url=https%253A%252F%252Fwww.google.com%252F, but it would be a different beast, with a dedicated power supply and a questionable price.