Hi all,
I'm building a thermocouple interface based on AD8237
http://www.analog.com/static/imported-files/data_sheets/AD8237.pdf instrumentation amplifier - my schematics is below. Yesterday I had an opportunity to use a 7 digit multimeter to test the performance of my design. Could someone go thru the process I've used and comment on that was I doing wrong?
I can't interpret the results I get - I'm getting somewhat 65uV error over the range of my interest (0mv - 45mv). This is within the spec of instamp, but very on the edge really.
I've measured offset voltage by shorting the inputs of instamp. By subtracting reference voltage from the output and dividing it by gain of 101 I get an offset of 20uV which is totally ok - distribution graph has a big peak around that area.
Since I don't have an accurately controlled oven to test it with a thermocouple, I've made myself a test rig to emulate the thermocouple - I've hooked up a 1.2V reference and a resistor divider out of 100k resistor and a 10k pot, i'm powering it from the battery. I went and made 45 measurements - in average they show somewhat ~65uV error.
Max bias current for this instamp is 650pA -- this can't be causing additional 45uV of offset.

What could be causing this error? Is my setup wrong? What could I change to test it another way?