Per request we've created this, hopefully it's not repeating something we've shown elsewhere.
This is more about LCR meter confidence and certainly not metrology grade, rather than resorting to expensive calibration which is necessary for actual client use, and will be applied when necessary.
It's about creating a direct BNC connection for reference components (mostly capacitors at this time) to a quality bench LCR meter. The general idea is to remove most of the uncertainty in LCR measurements attributed to the clips, cables, connectors and fixtures. Our goal was/is to show repeatability over time that will allow more confidence in LCR instrument behavior for our home lab instruments (Hioki IM3536 and Tonghui TH2830).
The idea starts with using 4 direct BNC connectors to the LCR meter Lc, Lp, Hp, and Hc terminals. A 3D printed container box with separate front and back cover and a rectangular "tube" to house the DUT within. The covers are attached with 3mm screws and all the "insides" of the covers and tubes covered with copper foil to create a somewhat Faraday cage for the DUT to reside within. This provides an effective E-Field shield which has shown to eliminate the effects of local distributes such as hands moving around the DUT when setting the LCR meter parameters, it's not an H-Field shield tho!!
To facilitate good Open and Short Calibrations, we created a dedicated Open and Short fixture as well. The Short uses a thick continuous copper shunt across the 4 BNC terminals.
Anyway, here's the actual fixtures showing the Open, Short, and various reference capacitors. All the caps are select mica except the 0.1uF and 1uF which are select PP film types. The special purple 10nF has a 5kΩ thermistor attached to the Soviet metal case for remote temp readings.
Best