EEVblog Electronics Community Forum
Electronics => Projects, Designs, and Technical Stuff => Topic started by: Liv on March 05, 2015, 01:27:03 pm
-
Interpolating reciprocal frequency counter FC-510
Features:
• reciprocal counting + interpolation (3-rd generation by Pendulum classification)
• main input frequency range: 0.1 Hz…100 MHz
• period range : 10 ns…10 s
• measured single pulse duration: 100 ns…10 s
• measured duty cycle : 0.0000001…0.9999999
• measured RPM: 6…999999 rpm.
• measuring interval: 1 ms…10 s
• analog interpolator resolution: 1 ns
• average count: 1…100
• IF range: ±99.9999 MHz
• math: Fmin, Fmax, dF
• hold: yes
• programmed prescaler ratio: 1…1000
• prescaler frequency range: up to 2 GHz
• input impedance: 1 MOhm || 30 pF
• input coupling: DC+AC, AC
• input attenuator: 1:1, 1:10
• sensitivity: 100 mV (up to 50 MHz)
• max input voltage: ±50 V
• comparator threshold: –0.5…+0.5 V; –5…+5 V
• optional LPF: 100 kHz
• reference: TCXO, ±2.5 ppm (-20…+70°C)
-
Some photos:
-
+1. I made this frequency meter about a year ago. Works like a charm.
For clocking I used a standard 10Mhz VTCXO with a trimmer resistor for fine freq. adjustment. Calibrated all thing by 1PPS from a GPS receiver.
-
Firmware version 2.x supports 9-digit frequency display (for 1601 and 1602 LCDs) and level meter in second line (for 1602).
Resolution demonstration (sorry for unintelligible language, I will try to fix):
video (https://www.youtube.com/watch?v=VNKCnef-ynU)
-
BTW, you forgot to mention that it is a _reciprocal_ counter, not a direct count one.
It makes a huge difference from specs perspective.
-
Thank you, I made correction of features list.
-
Don't forget the topic name, it must be eye-catching! :)
-
OK, fixed.