I have 2 HP8904A units.


float measurements taken with 34401a, 10mohm input impedance setting
Supposedly good unit:
channel 1 to earth : 31k (functional light)
channel 2 to earth ; 1k (functional light)

supposedly bad unit (i think it just has a broken indicator light for float)
channel 1 to earth: 45k (good light)
channel 2 to earth : 16k (non functional light)


I hooked up a DC supply to the float control relay and found that if I put enough DC into it I can get the impedance between chassis ground and the supposedly floating thing to read High Z on my 10Mohm input z ohm meter.

Do these units both have faulty relays/relay controllers... or is floating >1Kohm considered good ? I figured float ment broken circuit, so a DC ohm meter should read infinity...
« Last Edit: September 12, 2015, 09:16:16 pm by sarepairman2 »
 

Offline sarepairman2Topic starter

  • Frequent Contributor
  • **
  • Posts: 480
  • Country: 00
Re: hp8904a function generator floating impedance?
« Reply #1 on: September 12, 2015, 09:06:56 pm »
and when its not floating the impedance to earth ground/chassis is 50 ohms as it should be.


SO WHAT is considered a reasonable DC impedance for a floating instrument?

do these units have a common problem of the relay supply rail current being prone to fuck up?
« Last Edit: September 12, 2015, 09:17:46 pm by sarepairman2 »
 

Online tggzzz

  • Super Contributor
  • ***
  • Posts: 20580
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Re: hp8904a function generator floating impedance?
« Reply #2 on: September 12, 2015, 09:32:37 pm »
It isn't clear, but are you confusing "floating" with "isolated"?
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 

Offline sarepairman2Topic starter

  • Frequent Contributor
  • **
  • Posts: 480
  • Country: 00
Re: hp8904a function generator floating impedance?
« Reply #3 on: September 12, 2015, 09:54:35 pm »
It isn't clear, but are you confusing "floating" with "isolated"?

eh, the front chassis says "float" and it has a little light that turns on when its floating. '

 

Offline sarepairman2Topic starter

  • Frequent Contributor
  • **
  • Posts: 480
  • Country: 00
Re: hp8904a function generator floating impedance?
« Reply #4 on: September 12, 2015, 10:04:05 pm »
basically if the float is off the Low terminal center conductor reads like 0.5 ohms to chassis ground (shorted). when float is on it reads whatever I wrote (both my instruments are dual channel).

the impedance between high and low center pins in float mode is ~50 ohms and I verified that it works as a fully differential amplifier output.

im making a seperate thread for signal integrity of this instrument!