I have worked with the OP to figure out how the calibration constants are stored in the EEPROM and how they are used. Here is a summary, see attached for detailed notes.
High pressure:
0170 ADC value at 0 PSI
0172 ADC value at 100 PSI
Low pressure:
0194 ADC value at 0cmH2O
0196 ADC value at 120cmH2O
019a ADC value at -120cmH2O
Barometer:
0230 ADC value at 10.000inHg
0232 ADC value at 29.920inHg
All values are in 16-bit BE. The above sensors are all linear.
High flow:
0110 ADC value at 0LPM
0112 ADC value at 240LPM
The sensor is a hot-wire anemometer and its response is nonlinear.
Piecewise linear interpolation is used with a linearisation table at 0400-05ff.
This is a 256-entry table of values in units of 8192nds of the total output range (0-240000, units of 0.001LPM). See the notes for more detail.
Low flow:
0136 ADC value at 0LPM
0138 ADC value at 25LPM
The sensor is a hot-wire anemometer and its response is nonlinear.
Piecewise linear interpolation is used with a linearisation table at 0600-07ff.
This is a 256-entry table of values in units of 8192nds of the total output range (0-25000, units of 0.001LPM). See the notes for more detail.