Author Topic: Is 10x really the optimal attenuation for a passive oscilloscope probe?  (Read 1245 times)

0 Members and 1 Guest are viewing this topic.

Offline rs20Topic starter

  • Super Contributor
  • ***
  • Posts: 2318
  • Country: au
As Dave explained in one of his videos, IIRC, 10x probes have 10x attenuation for a couple of reasons:
-- To protect the scopes from high voltages (maybe?)
-- But mostly as a consequence of the passive probe cable being a deliberately highly lossy transmission line, because a lossless transmission line can't be built with a 1 megaohm characteristic impedance, a 50 ohm lossless transmission line would hopeless load down the circuit under test, etc, etc.

Meanwhile, oscilloscope manufacturers are pushing the limits of how low they can push the noise floor of their instruments, advertising ever lower voltages/division. But we're throwing away a factor of 10 most of the time!

So anyway, there's clearly a compromise in probe design here between A) load on device under test, B) reflections resulting from impedance mismatches at the ends of the probe screwing up the signal, and C) loss of SNR from the attenuation of the signal. And we've long settled on the 10x passive probe we all know and love today.

My question is this: is it really true that 10x is what 6-fingered aliens would conclude is the optimal compromise? Or are 10x probes a holdover from when scopes were analog and the attenuation had to be taken care of via mental arithmetic (for which dividing/multiplying by 10 is obviously easiest for us 10-fingered humans)?
« Last Edit: June 05, 2023, 11:25:46 am by rs20 »
 

Offline tggzzz

  • Super Contributor
  • ***
  • Posts: 19470
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
There are many many many different types of probes, each of which interacts with your circuit in different ways and has different characteristics. FFI see the references in https://entertaininghacks.wordpress.com/scope-probe-reference-material/

For even more information, see Tektronix and HP literature from the 60s.
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 
The following users thanked this post: Alex Eisenhut, ROT

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5884
  • Country: ca
normally i use x10 probes ....  and i use 100x probes who give much higher  impedance load  ....   but in very sensitive situations  you may need active probes,  but its another story


Some x100 probes can go up to 2.5 kv  on some models ...


Normally an x10 probe have x1 or x10 selection,  the important thing is to do the probe compensation adjustment

IE to have very square rising edges on the square wave output you have on the scope  ...
 

Offline Marco

  • Super Contributor
  • ***
  • Posts: 6719
  • Country: nl
I think it's still the best compromise. That said, 200 MHz range active 1x probes should have been commodity items by now.
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16607
  • Country: us
  • DavidH
-- To protect the scopes from high voltages (maybe?)

Probes which include shunt resistors can do that, however most x10 and x1000 probes do not and rely on the oscilloscope's 1 megohm input termination, so usually the oscilloscope needs to be able to handle full voltage at the probe tip whether the probe is attenuating or not.

Quote
-- But mostly as a consequence of the passive probe cable being a deliberately highly lossy transmission line, because a lossless transmission line can't be built with a 1 megaohm characteristic impedance, a 50 ohm lossless transmission line would hopeless load down the circuit under test, etc, etc.

The transmission line is made lossy to suppress reflections.  You can make a high impedance passive probe with 50 ohm cable, however it will be limited to accurately reproducing lower frequencies because of reflections in the cable.

The main reason attenuating probes are used is to lower the capacitance at the probe tip.

Quote
Meanwhile, oscilloscope manufacturers are pushing the limits of how low they can push the noise floor of their instruments, advertising ever lower voltages/division. But we're throwing away a factor of 10 most of the time!

Manufacturers like to advertise lower noise, however semiconductor physics has not changed.  If they are using JFET or MOSFET input stages, then their input noise is no better than oscilloscopes from the 1970s that used JFETs.  Usually noise is worse because they are using either a MOSFET input stage, or a higher frequency transistor so that the oscilloscope can be "upgraded" without changing the hardware.

Modern (since the 1980s) oscilloscopes tend to use dual path input buffers with a second low frequency input path which itself has attenuation at its input to isolate its input capacitance and increase its input range, but this also increases input noise at low frequencies.  Oscilloscopes are not designed to be low noise or they would not use this circuit topology.

My favorite oscilloscope inputs are high voltage (1) differential which inherently have like 5 times the input noise because of differential operation and the bootstrapping circuits needed for a wide input range.  They have visibly more noise unless bandwidth is limited to 5 MHz, but it is hardly ever a concern and the noise rejection of having a differential input often makes up for it anyway.

Quote
My question is this: is it really true that 10x is what 6-fingered aliens would conclude is the optimal compromise? Or are 10x probes a holdover from when scopes were analog and the attenuation had to be taken care of via mental arithmetic (for which dividing/multiplying by 10 is obviously easiest for us 10-fingered humans)?

10x is not far from optimal.  (2) 100x probes do not have significantly lower input capacitance.  5x probes used to be available but they were intended for trigger inputs on oscilloscopes and frequency counters where signal fidelity is less important than increased sensitivity, and even with a 5x probe on an oscilloscope, that would only double the sensitivity.  If you want lower noise measurements, then there are other ways to achieve it.

(1) High voltage differential in this case means +/-10 volts, which is provided by bootstrapping the input stage.  Old oscilloscope inputs typically operate with input signal range of +/-250 millivolts.  Modern cost reduced oscilloscopes operate with an input range 10 times higher without bootstrapping which has the disadvantages of requiring higher slew rates and creating more distortion, but it saves by having one less input attenuator.

(2) If we used base-8 arithmetic, then our probes would be 8x for convenience.  If we used base-12 arithmetic, then our probes would be 12x for convenience.  There is no application where the difference in noise would matter between 8x, 10x, and 12x probes.
« Last Edit: June 05, 2023, 05:10:17 pm by David Hess »
 
The following users thanked this post: eugene

Offline Hydron

  • Frequent Contributor
  • **
  • Posts: 985
  • Country: gb
PMK make some 20x probes as a compromise to get lower input C than a 10x probe (5.6pF), but higher signal level than a 100x. They are unfortunately annoyingly expensive and hard to get hold of. Note that a few manufacturers rebadge PMK probes - Keysight carry the 20x one I believe.
« Last Edit: June 05, 2023, 09:18:31 pm by Hydron »
 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 4525
  • Country: au
    • send complaints here
Passive probes, tradeoff of low capacitance vs high impedance at DC
10x for a 1M ohm input is a good balance and from history/momentum, just as 50 ohm. The "ideal" is probably a little off from there but it's close enough not to matter or to break the backwards compatibility/practice.
 

Offline tggzzz

  • Super Contributor
  • ***
  • Posts: 19470
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Passive probes, tradeoff of low capacitance vs high impedance at DC
10x for a 1M ohm input is a good balance and from history/momentum, just as 50 ohm. The "ideal" is probably a little off from there but it's close enough not to matter or to break the backwards compatibility/practice.

The "at DC" is an important caveat.

In practice the misnamed "high" impedance *10 probes are lower impedance than "low" impedance *10 Z0 resistive divider probes.
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 

Offline nigelwright7557

  • Frequent Contributor
  • **
  • Posts: 689
  • Country: gb
    • Electronic controls
A x1 probe with a high impedance signal is loaded down by 10pf input capacitance.
A x10 probe has a capacitor across the potential divider to compensate and so doesn't have the same losses.

 
The following users thanked this post: ozkarah

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16607
  • Country: us
  • DavidH
A x1 probe with a high impedance signal is loaded down by 10pf input capacitance.

A x1 probe has more like 200 picofarads of input capacitance.  It will depend on the length of the x1 probe's cable.

 

Offline Someone

  • Super Contributor
  • ***
  • Posts: 4525
  • Country: au
    • send complaints here
Passive probes, tradeoff of low capacitance vs high impedance at DC
10x for a 1M ohm input is a good balance and from history/momentum, just as 50 ohm. The "ideal" is probably a little off from there but it's close enough not to matter or to break the backwards compatibility/practice.
The "at DC" is an important caveat.

In practice the misnamed "high" impedance *10 probes are lower impedance than "low" impedance *10 Z0 resistive divider probes.
Pointing out the need for accuracy, and then immediately leaving a comment about impedance at a non-specific frequency  :palm:

Identification of the probes termination impedances and ratios are left as an exercise for the reader...
 

Online Alex Eisenhut

  • Super Contributor
  • ***
  • Posts: 3336
  • Country: ca
  • Place text here.
You can read the Tektronix Concept Series on the subject
https://w140.com/tekwiki/images/6/62/062-1146-00.pdf
Hoarder of 8-bit Commodore relics and 1960s Tektronix 500-series stuff. Unconventional interior decorator.
 