Electronics > Projects, Designs, and Technical Stuff
Keithley 2400 SourceMeter Quick Test (Help Needed!)
Hugoneus:
Dear members,
Does any of you have a Keithley 2400 SourceMeter handy?
If so, can you do quick output short measurements where you directly short the output terminals together without using a cable? What resistance does it report to you? (You have to make a very low-resistance short of course).
I would very much appreciate this.
mickpah:
Cf
D
Hugoneus:
--- Quote from: mickpah on December 03, 2012, 04:19:34 am --- Cf
D
--- End quote ---
Hmmm... That is not as helpful as I thought it would be! :-//
free_electron:
What are you after ? Simply turning on ohms meter function uses regular two wire sensing with these machines. You need to go in the menu and explicitly turn on 4 wire sensing. Then you can measure low stuff.
Hugoneus:
--- Quote from: free_electron on December 03, 2012, 05:23:31 am ---What are you after ? Simply turning on ohms meter function uses regular two wire sensing with these machines. You need to go in the menu and explicitly turn on 4 wire sensing. Then you can measure low stuff.
--- End quote ---
The problem is as follows:
I set the instrument to 1A and connect a precision 1-Ohm resistor at the output. I use the Measure-V function and it reports 1.3V. I put the multimeter across the unit's output terminals (right at the instrument) and I read 1.01V.
I would like to know if this is normal for this instrument to have an internal resistance of 0.3-Ohms... It seems rather high.