Electronics > Projects, Designs, and Technical Stuff
Keithley_DMM7510 burden voltage
(1/2) > >>
brumbarchris:
I am just looking in the datasheet of the Keithley DMM7510 (attached only the two relevant pages) and in the table on page 10 a 200mV burden voltage is indicated for the 100mA range. This would indicate a shunt resistance of some 2 Ohm. However, in the last table on page 11, an Effective Internal Sunt Value for the same range is indicated as 0.1Ohm. I am not sure I understand the discrepancy, can somebody explain?

Best regards,
Cristian
Kleinstein:
The first 2 tables with the higher drops seem to include the drop on the fuse and switching part. This is especially obvious for the higher currents. Normally one has to keep the drop at the shunt rather small at something like 3 or 10 A to limit self heating.

The last table to a large part looks like the actual shunt resistor used. However the value of 0.1 Ohms still looks low for the 100 mA range. 1 Ohms would make more sense  - so maybe an additional misprint there.

The low current ranges(<=10 mA)  seem to use a trans-impedance amplifier and thus way get so low voltage drops.
tooki:
For what it’s worth, in my 6.5 digit Keithley 2015, it uses a 0.1 Ohm shunt for the 100mA range, with a max burden voltage of 30mV, according to the specs in the manual.

So the 7510 spec doesn’t make obvious sense to me, either.
Kleinstein:
The Keithley 2015 like the 2000 is more like a low cost version with few shunts. So they use the same shunt for different ranges, compromising performance. So the 100 mA range would a kind of scaled up 1 A range.

The DMM7510 is a higher grade instrument, and likely uses separate shunts, expect for likely 1 A and 3 A. The specs suggest a separate shunt for the 100 mA range.
HighVoltage:
I just did a quick test on my DMM7510


mA range
1.1 mA, 11 mV, ~ 10 Ohm
10 mA, 12.7 mV, ~ 1.2 Ohm
100 mA, 126 mV ~ 1.2 Ohm

A range
0.2 A, 64 mV ~ 0.3 Ohm
1A, 0.3 V ~ 0.3 Ohm

So, at least 3 different shunt resistors.

Navigation
Message Index
Next page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod